Greetings,
I am searching here in an attempt to remedy a failure I get when performing the comprehensive low level calibration.
I have already replaced the Electrolytic Capacitors and the self test passes with no errors. I complete all of the steps of the calibrations once I enter the calibration due date. I get error +440 " Calibration not complete. Once I query the errors with SYST:ERR?
I get 200uA gain OOT +363. I have about 10 of these unit is house. I have only gotten 2 to pass the low level calibration successfully.
The one the passes has B17/A02. The two the fail so far have B16/A02. I think the issue is in the firmware.
I had planned to purchase blank AT27C040-90PU-ND, copy the firmware from my working unit and install in the other unit. BUT I think this will not w
We have a DATA man M-pro that will allow for copying the firmware, will I still need to perform the calibration after updating the firmware?
Any and all help would be appreciated