I started to get the itch to augment my meter collection with a 6.5 bench meter after hearing people here talk about getting good used gear from eBay. But after watching the prices for a bit, I came to thinking that if I bought a broken meter instead, not only was I buying a meter, I was buying hours of entertainment!
![Smiley :)](https://www.eevblog.com/forum/Smileys/default/smiley.gif)
Still, I was cautious and waited for something with a problem I thought I would be capable of fixing, and eventually found a Keithley 2000 which was failing self test 200.2. The service manual describes this as the as 200.1, but with a different resistor network switched in. I figured I could probably handle troubleshooting resistors!
The unit was amazingly clean inside. I feel guilty for leaving probe marks all over the place, hehe. I was able to piece together the schematic for the part of the board in question. As you can see, the 100 gain section is only producing a 90 gain, and the 1079 ohm resistor seems to be the likely candidate for being out of spec as a 1k in its place would produce the correct gain.
I took a picture of that part of the board, but sadly my camera (and photo skills) are not up to the task of taking pictures where you can read the part numbers. U166 is an
AD711, U129 is a
DG211DY, R309 is labelled 1k00 and R310 is labelled 9k09.
I did some quick algebra and computed that a roughly 14k resistor in parallel with the 1079 would bring it down to 1000, which would fix the gain ratio. So I wrapped a 10k and a 4.7k together and held them in place... voila! Passes the self test
Some questions for you all:
- What function does R311 serve? (or R308 for that matter.)
- I haven't modified the board yet, so to get those resistance values, I measured various points on the board and then set up systems of non-linear equations and guessed values in excel until I got close to my measurements. Is there a better way?
- What causes a portion of a resistor pack to change like that? The simplest repair would be to solder a 14k between pins 9 and 6 of the resistor pack, but can I trust the 1079 resistor to be stable now? Or am I better off bypassing the resistor pack entirely? Getting stable resistors in a 99:1 ratio seems pretty tough, but I suspect 100:1 would do.
(schematic done with www.circuitlab.com)