Author Topic: Help Fluke 8845 can't read resistance value  (Read 862 times)

0 Members and 1 Guest are viewing this topic.

Offline qnhn90Topic starter

  • Contributor
  • Posts: 10
  • Country: vn
Help Fluke 8845 can't read resistance value
« on: January 13, 2021, 02:37:44 am »
hi all
i have 1 ea fluke 8845 can't read resistance value.
Do you have a way to handle it thoroughly?
I can only measure up to 170k ohm.
and then other higher values ​​cannot be measured.
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 7994
  • Country: us
Re: Help Fluke 8845 can't read resistance value
« Reply #1 on: January 13, 2021, 02:52:25 am »
Are you saying that it reads **accurately** up to 170K, but then displays 'OVERLOAD' for anything higher than that?  Or is it inaccurate?
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline qnhn90Topic starter

  • Contributor
  • Posts: 10
  • Country: vn
Re: Help Fluke 8845 can't read resistance value
« Reply #2 on: January 13, 2021, 09:54:05 am »
If the value is higher, it is still only displayed at 170k ohm. no more.
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 14648
  • Country: de
Re: Help Fluke 8845 can't read resistance value
« Reply #3 on: January 13, 2021, 10:00:30 am »
Is the 170 K upper limit fixed, independent of the range (e.g. 2 M or 20 M range) ?

It could be something like the Ohms current source not able to provide a higher voltage or some additional resistance across the inputs (e.g. a bad MOV, bad Ohms source protection circuit, dirty board (leaking cap, fried bug).

A test worth while may be checking if the voltage reading input is also high impedance as it should. One can do this by charging a small cap (like 1-10 n) to some 10 V and than check how fast the capacitors discharges when measuring the voltage.
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 7994
  • Country: us
Re: Help Fluke 8845 can't read resistance value
« Reply #4 on: January 13, 2021, 02:48:17 pm »
If the value is higher, it is still only displayed at 170k ohm. no more.

OK, I assume this is on the 1 megohm range.  First, try manually putting it in the 10 megohm range and see what the reading is for a larger resistor.  Then use another meter with at least a 10M input impedance and measure the voltage output of the 8845A in the 1M and 10M resistance ranges.

Have you had the meter for a while and this all suddenly just happened or did you get it with this problem?
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline wn1fju

  • Frequent Contributor
  • **
  • Posts: 569
  • Country: us
Re: Help Fluke 8845 can't read resistance value
« Reply #5 on: January 13, 2021, 06:19:16 pm »
I've had several "broken" meters (not this particular Fluke model) that behaved strangely.  Finding nothing overtly wrong with the circuitry, I ran through the calibration
procedure and things then worked fine.  It seems that a lot of times, the calibration constants get corrupted and the meter will produce erroneous results until they are
overwritten with values that make sense.
 

Offline mzacharias

  • Frequent Contributor
  • **
  • Posts: 710
  • Country: us
Re: Help Fluke 8845 can't read resistance value
« Reply #6 on: January 14, 2021, 12:21:29 am »
I would use another multimeter to measure the input resistance of the Fluke on its DC volts mode.

I'm betting a damaged part in the front end, coincidentally measuring - wait for it - 170K
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf