Hi,
For those playing along at home, here is the part of the schematic with T3:
You can disconnect the parts that I have circled in Red. The unit will work without the fan or the X cap C26.
You can isolate the transformer by removing the full-bridge rectifier U1 and disconnecting one end of the rectifiers CR6-9.
If the fuse still blows with only the transformer connected, it is possible the transformer has a defect called a shorted-turn. This means the resistance of the windings will read correctly or close to correctly, but the shorted turn will cause excessive current to flow in the primary, blowing the fuse F2.
Regards,
Jay_Diddy_B