Author Topic: Holt Instrument Laboratories Model 611 Voltage Calibrator  (Read 10818 times)

0 Members and 1 Guest are viewing this topic.

Offline vsmithTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Holt Instrument Laboratories Model 611 Voltage Calibrator
« on: January 06, 2024, 10:38:52 pm »
Hi All

I have a Holt model 611 Voltage Calibrator, for which I have no information whatsoever. It is an AC calibrator with built in 400 Hz capability and an external signal input accepting 50-1000 Hz. It is a mid 60s instrument, judging form the components, manufacture style and nomenclature (cycles!), with a mainframe and plug in card cage type form.

It powers up, but the output does not agree with the switch settings (3v is actually 2.28 v for example), so it appears to be repairable but without a manual or schematic, a pretty difficult undertaking!

Can anyone help me with documentation? Does anyone on the forum have or ever worked with such a beast? What even are it's capabilities (I like the promise of 1000V output!)

photo of the panel attached.

Thanks!

Vic
 

Offline TheDefpom

  • Frequent Contributor
  • **
  • Posts: 725
  • Country: nz
  • YouTuber Nerd - I Fix Stuff
    • The Defpom's Channel
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #1 on: January 08, 2024, 11:23:20 pm »
Generally speaking, if the output is low it could be due to bad caps causing AC ripple which then effects the output control... 1st thing is to check the power supplies and make sure the ripple is not bad on the DC rails.
Cheers Scott

Check out my Electronics Repair, Mailbag, or Review Videos at https://www.youtube.com/TheDefpom
 

Offline vsmithTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #2 on: January 10, 2024, 07:56:32 pm »
Thanks for your reply, that is good advice, and will be my next step. My difficulty is even finding the "rails" as the power supplies are not the usual "full wave bridge-3 terminal regulator" types used in later era equipment. Everything is discrete (no op-amps that I can see, yet) There are three large electrolytic caps, and definitely two rectifier (heat sinked) diodes in a full wave - center tapped transformer secondary type configuration, but I have not found the return path (the "ground" point) yet. I have checked those caps and they are close to their nominal values. I will have to do some more detailed sleuthing. Really could use a schematic!!
 

Offline vsmithTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #3 on: January 11, 2024, 01:48:00 pm »
Just an update.

Tracing out the circuit (and I did find the "ground" return for the full wave rectifier) it is a common positive, regulated negative power supply. Center tapped transformer secondary, two rectifiers, a 1.5 Ohm series resistor, 2 parallel 1800 uF input filters, another 1.5 Ohm resistor, another 1800 uF cap (raw unregulated voltage aprox 30V) and then a 20V supply built on two zeners and a series of PNP (Germanium) transistors, with the final Pass element heat sinked. It appears as though the 20V is used for all the circuitry. All the filter caps measure "good" both with an ESR type meter and in circuit. The final regulated output is "quiet" (4mv ripple) and stable (as expected, some warm up drift). It is adjustable with a trimpot, and was noisy until I exercised it, leaving it set at 20.0V. I assume that further downstream regulation is used on each circuit card (there are 4) to produce a stable 400Hz sine wave for the output. More on that to come.
 

Offline vsmithTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #4 on: January 11, 2024, 02:16:50 pm »
some photos of the insides
 
The following users thanked this post: ch_scr

Offline vsmithTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #5 on: January 11, 2024, 02:18:15 pm »
and more photos showing detail - in the photo of the card cage, the second card in has a nuvistor!
 
The following users thanked this post: ch_scr

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 6206
  • Country: ca
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #6 on: January 14, 2024, 02:56:01 am »
if its a repair  you are not in the good section

secondo    your lytics seems bad just looking at them ???    or some crust on them
 

Offline Haenk

  • Super Contributor
  • ***
  • Posts: 1161
  • Country: de
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #7 on: January 14, 2024, 10:52:29 am »
Ouch, those caps really look like they are returning from the undead soon.
Before anything else, these certainly have to be replaced, they look beyond awful.
 

Offline factory

  • Super Contributor
  • ***
  • Posts: 2959
  • Country: gb
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #8 on: January 14, 2024, 12:52:50 pm »
It's only the sleeving on those capacitors that looks bad, they are probably fine, with the exception of the smallest diameter ones, that go high ESR or open for most brands.

David
 

Offline vsmithTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: us
Re: Holt Instrument Laboratories Model 611 Voltage Calibrator
« Reply #9 on: January 14, 2024, 03:11:00 pm »
Thanks for those "tips" about the caps. Yes, they look "beyond awful" but yes, it is mostly external. The speckled yellowish axial caps are Sprague "TE" types, they have a soft poly sleeving which seems to have reacted with the storage conditions (garage) over a decade or so. It can easily be scraped away and is a waxy sort of substance. The circuit boards are all conformal coated and these caps may have been coated with something as well, not sure. At any rate, I have actually checked each and every one (in circuit) using a handy capacitance/esr meter. I found only one that definitely needed replacement - it was a small (10 uf at 25v) one that had gone open.

I have reposted this topic in the Repairs section, I agree that it probably belongs there, although the unit is a piece of calibration equipment. Thank you all for your comments. I really hope to luck out and find someone who has a schematic or even better a manual for this.

In closing, at this point I have a working calibrator, but the output is not stable. I have traced out a lot of the signal flow from card to card, but don't really understand how the output level is regulated (I have found a "black box" on one card which is completely unlabeled but I suspect is a thermo-electric ac-dc converter for the control loop) but I really need some sort of "how is it supposed to work, as designed" resource!
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf