Author Topic: Is this scope repairable?  (Read 22761 times)

0 Members and 1 Guest are viewing this topic.

Offline MediarockerTopic starter

  • Contributor
  • Posts: 46
  • Country: us
  • If it ain't broke, open it anyway!
Re: Is this scope repairable?
« Reply #25 on: October 28, 2012, 01:19:10 pm »
Try loading default calibration. But make sure you switched the switch on the back of the instrument.

Yes, I did that. But I'm trying to Self Calibrate it... is it just not capable of self calibration anymore?
 

Offline MediarockerTopic starter

  • Contributor
  • Posts: 46
  • Country: us
  • If it ain't broke, open it anyway!
Re: Is this scope repairable?
« Reply #26 on: October 29, 2012, 10:27:50 pm »
Any ideas anyone? I have a feeling channel 3 has a bad input.

But what would cause this during calibration?
 

Online Psi

  • Super Contributor
  • ***
  • Posts: 10151
  • Country: nz
Re: Is this scope repairable?
« Reply #27 on: October 29, 2012, 10:50:56 pm »
Any ideas anyone? I have a feeling channel 3 has a bad input.

Maybe CH3 has a fault which is blocking the delay system from working on any channel.

Are they modular at all? can you pull out the channel 3 pcb ?
Greek letter 'Psi' (not Pounds per Square Inch)
 

Offline MediarockerTopic starter

  • Contributor
  • Posts: 46
  • Country: us
  • If it ain't broke, open it anyway!
Re: Is this scope repairable?
« Reply #28 on: October 30, 2012, 07:16:45 am »
Any ideas anyone? I have a feeling channel 3 has a bad input.

Maybe CH3 has a fault which is blocking the delay system from working on any channel.

Are they modular at all? can you pull out the channel 3 pcb ?

Unfortunately this is the board here. I would have to think that everything having to do with the channels would be located in those silver boxes.



The only issue I'm having is with Calibrating the delay trigger and on channel 3 it seems as if the input has been zapped..
 