Author Topic: Tektronix 7854 horizontal adjustment  (Read 696 times)

0 Members and 3 Guests are viewing this topic.

Offline little nellieTopic starter

  • Newbie
  • Posts: 1
  • Country: us
Tektronix 7854 horizontal adjustment
« on: November 23, 2019, 03:46:13 am »
Hi, is there a way to adjust the horizontal position in the 'store' mode? I am able to adjust the horizontal position in the 'scope' mode with the timebase. But when switched to 'store' the display is shifted left about 1/8", unable to locate any knobs for adjustment in that mode.  thanks
« Last Edit: November 23, 2019, 03:54:20 am by little nellie »
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 16969
  • Country: us
  • DavidH
Re: Tektronix 7854 horizontal adjustment
« Reply #1 on: November 23, 2019, 02:44:02 pm »
The service manual includes all of the details for the rather involved display calibration procedure.  There are trimmer potentiometers to adjust position and scaling of both the analog and storage display elements so that they align with the graticule and align with each other.
 