Many thanks Tony_G for all information,
I was to quick and done my first measurement wrong with input set to 1Meg as you mentioned, but also realized my error in the meantime. Setting it to 50ohm I get also a rise time around 300ps and a signal of about 350mVpp what is expected. So, this is good but still my Trigger calibration does not pass
![Sad :(](https://www.eevblog.com/forum/Smileys/default/xsad.gif.pagespeed.ic.L3FGyzQrjB.png)
I have tried to do the Trigger system performance tests from the service manual with a signal generator (100,500,1000Mhz) and all pass. So, this is all strange. If have also measured the trigger delay of my scope AUX out set to Trigger out and it measures 25ns, can you maybe check this also?
I even recorded the whole calibration procedure. Put a low load 1.5Ghz active probe to look what goes on during calibration. I was quite surprised that quite some different waveforms come out when the scope performs the calibration. Other scope was set to 500mV/4ms per division, trigger set to -140mV and sensitive to both edges, persistence off. What is maybe strange that to the end of calibration nothing happens it’s all quiet for quite some minutes...
Video of CH2 calibration (CH1 is longer but as both fail i did onlly the video of CH2)
https://drive.google.com/open?id=0B-jLzsxL7QoAanpFQUhrWUV3ZVE