Hello everyone

I'm a newcomer to electronics and by luck I recently acquired an old 54502A 400MHz digitizing oscilloscope that was being thrown away: https://ibb.co/M9bfnMp

Unfortunately the cathode ray tube display is broken. When I use the 'self-test' function the entire display is supposed to light up, yet only part of it down the centre does: https://ibb.co/hKXV2F2

If it's just a case of replacing a component I'll try and repair it but if you think the display itself is busted I'll take it apart for parts.

Thanks for the help
« Last Edit: February 23, 2020, 09:14:13 pm by brych34 »
 

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 790
  • Country: us
Re: Agilent 54502A CRT display broken
« Reply #1 on: September 27, 2019, 01:38:36 pm »
That CRT display looks great. Your problem is likely in the horizontal deflection. Likely the vertical mounted board to the left of the CRT. Unfortunately I don't think there is a schematic for it.
 
The following users thanked this post: brych34

Offline brych34Topic starter

  • Contributor
  • Posts: 15
  • Country: wales
Re: Agilent 54502A CRT display broken
« Reply #2 on: September 27, 2019, 04:16:21 pm »
Thanks - I'll have a look at it tonight after I discharge the display.
 

Offline TurboTom

  • Super Contributor
  • ***
  • Posts: 1406
  • Country: de
Re: Agilent 54502A CRT display broken
« Reply #3 on: September 27, 2019, 08:00:09 pm »
These monochrome CRT displays usually run their horizontal deflection through a quite substantial bipolar electrolytic capacitor of a few microfarads. Over the decades, these caps will dry and ESR will increase until the horizontal deflection fails (almost) completely. You may want to search for such a capacitor on the CRT PCB and check its capacitance / ESR. Other components in its vicinity may fail as a consequence.
 
The following users thanked this post: fcb, brych34

Offline brych34Topic starter

  • Contributor
  • Posts: 15
  • Country: wales
Re: Agilent 54502A CRT display broken
« Reply #4 on: September 28, 2019, 12:02:00 pm »
I'll see if I can get my hands on an LCR meter. Thanks
 

Offline brych34Topic starter

  • Contributor
  • Posts: 15
  • Country: wales
Re: Agilent 54502A CRT display broken
« Reply #5 on: February 23, 2020, 09:11:38 pm »
It works now! I replaced the 3.3uF bipolar capacitor on the board to the left of the CRT with one from AliExpress:


As far as I can tell it works perfectly now:



Thank you!
 

Offline GreyWoolfe

  • Supporter
  • ****
  • Posts: 3652
  • Country: us
  • NW0LF
Re: [SOLVED] Agilent 54502A CRT display broken
« Reply #6 on: February 24, 2020, 01:39:49 pm »
 :-+ :-+
"Heaven has been described as the place that once you get there all the dogs you ever loved run up to greet you."
 
The following users thanked this post: brych34

Offline fcb

  • Super Contributor
  • ***
  • Posts: 2125
  • Country: gb
  • Test instrument designer/G1YWC
    • Electron Plus
Re: [SOLVED] Agilent 54502A CRT display broken
« Reply #7 on: February 24, 2020, 02:19:38 pm »
Excellent repair. These are great scopes - especially for free.

Remember they are equivalent time sampling at 400MHz. Realtime to 100MHz.  But still excellent!
https://electron.plus Power Analysers, VI Signature Testers, Voltage References, Picoammeters, Curve Tracers.
 
The following users thanked this post: brych34

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 790
  • Country: us
Re: [SOLVED] Agilent 54502A CRT display broken
« Reply #8 on: February 24, 2020, 09:21:09 pm »
Great Job!  I'll try to remember that fix. Lots of 54xxx scopes, logic analyzers, and the 53310A Modulation Domain Analyzer use that same CRT/board combo. How did you happen to find the cap???
 

Offline TK

  • Super Contributor
  • ***
  • Posts: 1722
  • Country: us
  • I am a Systems Analyst who plays with Electronics
Re: [SOLVED] Agilent 54502A CRT display broken
« Reply #9 on: February 24, 2020, 09:38:53 pm »
I had the same issue with an HP8714B and replaced the bipolar capacitor.  They come in a much smaller footprint now and it worked when I replaced it... but then later another magic smoke and the new capacitor was gone.  I asked the forum and I remember somebody recommended using a non electrolytic capacitor and that is what I did and it has been working for 3-4 years now.
 
The following users thanked this post: brych34

Offline brych34Topic starter

  • Contributor
  • Posts: 15
  • Country: wales
Re: [SOLVED] Agilent 54502A CRT display broken
« Reply #10 on: April 09, 2020, 04:14:05 pm »
How did you happen to find the cap???

Sorry for the late reply. I just searched for "3.3uF bipolar capacitor" on AliExpress and it led me to here: https://www.aliexpress.com/item/32594169451.html

So far it seems to be working well!
 
The following users thanked this post: mightyohm


Share me