I have now replaced the resistors to SMD type 0805 with max. 5ppm/K and it seems the circuit is now much more stable with ambient lab conditions. Otherwise noted in the datasheet I had to recalculate the resistor value (original 9k/0,1%) in the feedback path for selected references to match my 10.00x V requirements.
1: 6.84409V LM399H (National Semiconductor) --> 15k || 24k
2: 6.92224V LM399AH (National Semiconductor)
3: 6.88884V LM399H (Linear Technology)
4: 6.87789V LM399AH (National Semiconductor)
5: 6.88650V LM399AH (National Semiconductor)
6: 6.86366V LM399AH (National Semiconductor)
7: 6.91630V LM399AH (National Semiconductor) --> 15k || 22k
Number 1 is (still) running under lab conditions, the HP 34410A has warmed up for more then one week:
12.03.2013: 10.00340V / 10.00341V
13.03.2013: 10.00338V / 10.00339V
14.03.2013: 10.00337V / 10.00338V at morning & 10.00338V / 10.00339V at closing time
15.03.2013: 10.00337V / 10.00338V at morning
It is planed to run a "Burn-In" (168h @ 125°C), but that will take some preparing time. I've also ordered some crystal heaters, that could be mounted on top of the resistors:
http://www.kuhne-electronic.de/en/products/crystal-heater/qh-40-a.htmlBy now I'm fine with the results, but will watch out the further drift.
P.S. It's the nature of an engineer to ask what is possible, not was is required, to reach the limits and try to pass them. We search for absolut accuracy and are unhappy with relative accuracy, unhappy with physical limitations and defacto standards, knowing that we are living in a relative Einstein's universe, where constants are variable
![Wink ;)](https://www.eevblog.com/forum/Smileys/default/xwink.gif.pagespeed.ic.cldandycH0.png)