Hi Forum Members,

There seem to be as many ESR meter designs as people who use them. I have decided to share my ESR meter design with the group.
Strictly speaking this ESR meter does not measure ESR, it measures the impedance of the capacitor under test at 100kHz. For Capacitors larger than a few uF this is approximately equal to the ESR.

ESR meter do not need to be precise, it most instances a bad capacitor will have any ESR many times that of a good one.

When I was kid, radio manufacturers would sell transistor radios based on the number of transistor in them. This is a 5 transistor ESR meter. The only IC that used is a 5V regulator. The circuit should also work directly from 3 or 4 AA batteries without the regulator. The design should work with any general purpose transistors. The diodes in the meter circuit will effect the meter scale. Use small signal Schottky diodes such as 1N5711, common diodes like BAT54 have too much capacitance. OA91 or similar Ge diode should also work.

10th Anniversary Edition

I shared a modified version the ESR meter in this message:

https://www.eevblog.com/forum/projects/5-transistor-esr-meter-design/msg4608886/#msg4608886

Two additional transistors are used to detect if the resistance of the capacitor being tested is less than 25 \$\Omega\$  All the other features remain the same. The new design was generated in KICAD version 5.00





The maximum voltage on the capacitor under test is 100mV. This allows for in-circuit testing.

The design started with modelling the circuit in LTspice:


The building blocks are labelled.

This is the result of the LTspice modelling, it shows how the meter deflection changes with ESR.



This good because it gives an expanded scale.

From the LTspice model, I made the OrCad schematic:



And then the board layout, I used SMD parts, but thru hole parts could also be used.



The circuit was made from the above design and the components attached:



The meter circuit was hooked up. The meter deflection was measure using some low value resistor. Artwork for a new meter scale was drawn using Corel Draw. The meter scale was printed on Photo paper to be attached to the meter.


The final assembly was completed. Here is a view of the inside of the ESR  meter:



And the Outside:



If you build this, remember to discharge capacitors before using.
Current consumption is 26mA, that is about 25 hours of use.

Regards,

Jay_Diddy_B aka John
« Last Edit: December 30, 2022, 10:50:50 pm by Jay_Diddy_B »
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7549
  • Country: 00
  • +++ ATH1
Re: 5 Transistor ESR Meter Design
« Reply #1 on: December 14, 2012, 04:15:57 am »
Thanks for sharing !  :-+

Any chance you could share the LTSpice file by attaching it to your 1st post ? Zip it 1st then attach.

Calling Kripton2035 ...   :)
« Last Edit: December 14, 2012, 04:18:41 am by BravoV »
 
The following users thanked this post: m3vuv, bmdaly, Synthtech

Offline Jay_Diddy_BTopic starter

  • Super Contributor
  • ***
  • Posts: 2765
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #2 on: December 14, 2012, 04:39:14 am »
I have added the LTspice file to the original message.

To obtain the graph.

Run the LTspice file. The circuit will run 15 times.
This from this SPICE Directive:

.step param R LIST 0.0001 0.1 0.2 0.5 1 2 4 6 8 10 12 14 16 18 20

When the analysis is finished from the VIEW menu 'Spice Error Log'

Then right click in the Error log window.

Click plot .step'ed .meas data

The curve will then be displayed.


Jay_Diddy_B aka John

 
The following users thanked this post: m3vuv

Offline toli

  • Frequent Contributor
  • **
  • Posts: 313
  • Country: il
Re: 5 Transistor ESR Meter Design
« Reply #3 on: December 14, 2012, 07:42:57 am »
Very nice work  :-+

Would have built one myself, but came across this a couple of weeks ago. For that price I think I'll give that one a try. Some more info can be found here.
My DIY blog (mostly electronics/stereo related):
http://tolisdiy.com/
 

Offline Jay_Diddy_BTopic starter

  • Super Contributor
  • ***
  • Posts: 2765
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #4 on: December 14, 2012, 01:16:41 pm »
Toli,

I have had a look at the eBay offering. It looks a lot the AVR Transistor Tester that was started by Markus.

A very detailed description can be found here:

http://www.mikrocontroller.net/attachment/143988/TTester_096k.pdf

It is possible that the software has been expanded to add additional functions. If you search 'markus AVR transistor tester' and look at the images you will find a lot of variations of this design.

It doesn't look like the unit can make in-circuit measurements, but I might be wrong.

You will have to let us know how well this unit works. May be somebody else has one?

In my design, I wanted to see if I could do something useful with a few transistors.

Regards,

Jay_Diddy_B
 

Offline toli

  • Frequent Contributor
  • **
  • Posts: 313
  • Country: il
Re: 5 Transistor ESR Meter Design
« Reply #5 on: December 14, 2012, 01:56:45 pm »
1. It is similar but of a later version. You can read more in the link I added to diyAudio forum.
2. Says the maximum voltage in the capacitor/ESR measurement range is 300mV, so it should be good for in circuit measurements as well.
My DIY blog (mostly electronics/stereo related):
http://tolisdiy.com/
 

Offline cwalex

  • Frequent Contributor
  • **
  • Posts: 299
  • Country: au
Re: 5 Transistor ESR Meter Design
« Reply #6 on: December 15, 2012, 11:49:36 am »
That looks awesome! Really professional finish, looks like you used a cnc mill for a lot of it.  :-+
 

Offline Jay_Diddy_BTopic starter

  • Super Contributor
  • ***
  • Posts: 2765
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #7 on: December 15, 2012, 02:57:36 pm »
cwalex and the group,

The circuit board and the front panel were made with a LPKF Protomat C60.



Here are some of the capabilities of this awesome machine:

  • Minimum Track Width 4 mil (0.1 mm)
    Minimum Insulation Width 4 mil (0.1 mm)
    Minimum Drill Hole Size 8 mil (0.2 mm)
    Working Area 13.4" x 7.8" (340 mm x 240 mm)
    Resolution 0.3125 mil (7.937 µm)
    Repeatability +/- 0.2 mil (5 µm)
    Precision of Front-to-Back Alignment +/- 0.8 mil (20 µm)
    Spindle Motor 3 phase-motor, 5,000-60,000 rpm, software controlled
    Tool Change semi-automatic, software prompted
    Tool Collet 1/8" quick release collet
    Drilling Speed 90 strokes / minute
    Travel Speed 1.575" per second (40 mm/sec)
    Milling Depth Sensing coaxial mechanical sensing
    X/Y Positioning System stepper motors, precision lead screws
    ActiveCAM® anti-backlash Supermuts®
    X/Y Linear System precision linear bushings and shafts
    Z Drive electromagnetic with hydraulic damper
    Dimensions (w / h / d) 17" x 14" x 22.25" (430 x 355 x 565 mm)
    Weight 53 lb. (24 kg)
    Power Consumption 200 VA


I was extremely lucky to find this, complete with the tools, vacuum and software at a reasonable price.

After milling the board was tinned using MG Chemicals 'Liquid Tin'.

http://www.mgchemicals.com/products/prototyping-and-circuit-repair/prototyping/liquid-tin-421/


Jay_Diddy_B
« Last Edit: November 19, 2013, 08:44:57 pm by Jay_Diddy_B »
 

Offline FenderBender

  • Super Contributor
  • ***
  • Posts: 1115
  • Country: us
    • The Solid State Workshop
Re: 5 Transistor ESR Meter Design
« Reply #8 on: December 16, 2012, 06:35:08 pm »
Did you just use some precision resistors for setting the scale? How many did you use?
 

alm

  • Guest
Re: 5 Transistor ESR Meter Design
« Reply #9 on: December 16, 2012, 07:17:29 pm »
Only five transistors... plus a voltage regulator with more than a dozen transistors ;)
 

Offline Jay_Diddy_BTopic starter

  • Super Contributor
  • ***
  • Posts: 2765
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #10 on: December 16, 2012, 07:18:59 pm »
FenderBender & the group,


I would suggest that you get 5 x 1 Ohm and 4 x 10 Ohm  1% resistors

This lets you get the following values:

0.2 Ohm 5 x 1 Ohm in parallel

0.25 Ohm 4 x 1 Ohm in parallel

0.333 Ohm 3 x 1 Ohm in parallel

0.5 Ohm 2 x 1 Ohm in parallel

1 Ohm 1 x 1 Ohm

1.5 Ohm 1 x 1 Ohm + 2 x 1 Ohm in parallel

2 Ohm 2 X 1 Ohm in series

2.5 Ohm 4 x 10 Ohm in parallel

3 Ohm 3 x 1 Ohm in series

4 Ohm 4 x 1 Ohm in series

5 Ohm 2 x 10 Ohm in parallel

Then add 1 Ohm in series to get 6, 7, 8, 9 Ohms

10 Ohms one 10 Ohm resistor


I measured the deflection with the original meter scale. Plotted a graph in Excel and calculated the deflection angles for the marks on the meter.

Keep the leads short especially when measuring the lower values.

Does this make sense ?

After I had finished the ESR meter I checked the calibration with some electrolytic capacitors that had been measured with a HP 4275A LCR (Z) Meter. This LCR meter works at frequencies from 10 kHz to 10 MHz, has a basic accuracy of 0.1% and 5 1/2 Digit display. :-+

Jay Diddy B
 

Offline FenderBender

  • Super Contributor
  • ***
  • Posts: 1115
  • Country: us
    • The Solid State Workshop
Re: 5 Transistor ESR Meter Design
« Reply #11 on: December 16, 2012, 07:43:31 pm »
So you used the 50uA scale since it is a linear scale. You measured the total angle of that scale? And then you did some proportions or something?

Corel Draw makes this easy? I understand how you did the math. Not so sure how I would go about drawing an accurate scale. I have Gimp..not too fluent in it.
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16362
  • Country: za
Re: 5 Transistor ESR Meter Design
« Reply #12 on: December 16, 2012, 07:55:23 pm »
I did a similar thing with mine, just skipped the printing of the scale thing. Used white out to remove the original scale markings, then sat with some resistors and did the same plotting of where the pointer was deflected to, then wrote the values with a Rotring drafting pen. Not the neatest or the most accurate, but does the job.
 

Offline Jay_Diddy_BTopic starter

  • Super Contributor
  • ***
  • Posts: 2765
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #13 on: December 16, 2012, 08:22:15 pm »
FenderBender & the group,

I measured the meter that I had using digital caliper. In my case the meter was 90o total deflection. Corel Draw defines 0o as vertical. So the angles required are +/- 45o.

Here is an image of the scale under construction:



The construction lines were deleted when the scale was finished.

Corel Draw is probably not the best software to us for this, but it is what I had available.

There is some special software available for doing custom scales. Here is a link to one that get good reviews:

http://tonnesoftware.com/meter2.html

I believe the MeterBasic, free version, only supports linear scales. I could not justify the $ for one meter.

Jay Diddy B
« Last Edit: January 25, 2013, 09:54:39 pm by Jay_Diddy_B »
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 38641
  • Country: au
    • EEVblog
Re: 5 Transistor ESR Meter Design
« Reply #14 on: December 21, 2012, 03:42:24 am »
Old Skool, nice!  :-+

Dave.
 

Offline KD0CAC John

  • Frequent Contributor
  • **
  • Posts: 710
  • Country: us
Re: 5 Transistor ESR Meter Design
« Reply #15 on: December 21, 2012, 02:42:18 pm »
Both the ARRL [ American Radio Relay league ] books , handbook & antenna book ,  have CDs in then with lots of software , ment for ham radio electronic repair & building .
I would say free , except you have to buy the books :) 
 

Offline HardBoot

  • Regular Contributor
  • *
  • Posts: 160
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #16 on: December 23, 2012, 06:12:03 am »
Really cool gadget, wish I could buy something like that for a few bucks and instead of a meter output to a micro, then a device could monitor it's own caps.
 

Offline Jay_Diddy_BTopic starter

  • Super Contributor
  • ***
  • Posts: 2765
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #17 on: January 10, 2013, 09:34:38 pm »
Hi group,

Has anybody else built this ESR?

Would there be any interest in a thru-hole circuit board design?

Regards,


Jay_Diddy_B

 

Offline redlandman

  • Newbie
  • Posts: 1
Re: 5 Transistor ESR Meter Design
« Reply #18 on: January 11, 2013, 04:59:05 am »
Jay_Diddy_B

I really like the design. I'd like to build it on a single sided board with standard componenets. I've made a few PCB's, but havn't been able to use any of the PCB design tools. I think it would be a great project, as yours has very good low end resolution.

Redlandman
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7549
  • Country: 00
  • +++ ATH1
Re: 5 Transistor ESR Meter Design
« Reply #19 on: November 10, 2013, 03:48:52 pm »
Bumping an old thread when looking at various diy esr meter circuits, Jay_Diddy_B, is there any possibilities of modifying the circuit, while still using jelly bean components, but the output is for digital volt meter instead of analog one ? Wishing this will not make the circuit overly complicated compared to the original version.  :-[

The reason I asked is finding a good analog meter is not easy, or at least cheap here in my place, which defies the original purpose that this circuit supposed to be simple, easy to build & cheap ESR meter.  :-//

Hope I'm not asking too much.

Offline kripton2035

  • Super Contributor
  • ***
  • Posts: 2683
  • Country: fr
    • kripton2035 schematics repository
Re: 5 Transistor ESR Meter Design
« Reply #20 on: November 10, 2013, 04:00:53 pm »
I ended up the same way when I build my esr meter years ago.
The go meter didnt exists, I focused on the analog poptronix esr meter
and did not find nor have in a desk the analog meter
so I used a bargraph and a lm3916 and it still works fine !
today I could use more segments (2x10 bargraphs) or the nice 28 segments display I found on ebay.

http://www.barmeter.com/en/viewproduct.asp?ID=184&XiajiID=0
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7549
  • Country: 00
  • +++ ATH1
Re: 5 Transistor ESR Meter Design
« Reply #21 on: November 10, 2013, 04:08:24 pm »
Yeah, maybe for experience EEs or pros which probably have many old unused T&M gadgets that still using analog meter that easily pull one for use, but its quite a hard for enthusiasts like me which usually do not have this kind of old analog gears around.  :'(

But my wish still reasonable I think, especially nowdays a DMM is a electronics noobs must have tool.
« Last Edit: November 10, 2013, 04:11:28 pm by BravoV »
 

Offline dannyf

  • Super Contributor
  • ***
  • Posts: 8221
  • Country: 00
Re: 5 Transistor ESR Meter Design
« Reply #22 on: November 10, 2013, 04:20:27 pm »
Quote
but the output is for digital volt meter instead of analog one ?

Yeah. I made one, far simpler than the OP's, using logic gates + some diodes, with a voltage read out -> making it an adapter for a digital meter.

As to analog read-out, analog multi-meters are fairly cheap, either in store or on ebay.

The fundamental issue with something like this is that they are NOT ESR meters -> they are ac impedance measurement tools. So you have to run them at high frequency to sufficiently null the impact of the capacitance.
================================
https://dannyelectronics.wordpress.com/
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7549
  • Country: 00
  • +++ ATH1
Re: 5 Transistor ESR Meter Design
« Reply #23 on: November 10, 2013, 04:23:01 pm »
Quote
but the output is for digital volt meter instead of analog one ?

Yeah. I made one, far simpler than the OP's, using logic gates + some diodes, with a voltage read out -> making it an adapter for a digital meter.

Schematic ?

Offline Jay_Diddy_BTopic starter

  • Super Contributor
  • ***
  • Posts: 2765
  • Country: ca
Re: 5 Transistor ESR Meter Design
« Reply #24 on: November 10, 2013, 06:02:20 pm »
Hi,

Before I designed the 5 Transistor ESR meter, I designed and built an ESR meter using a digital panel meter.

The design is unusual in that it uses 10kHz, instead of the more usual 100kHz. By using a synchronous rectifier to measure the in-phase component of the capacitor impedance.

Here is the schematic:

Update - I have corrected the schematic




I have attached the schematic in pdf format so that it will be easier to read.

Jay_Diddy_B
« Last Edit: November 11, 2013, 02:03:21 am by Jay_Diddy_B »
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf