Author Topic: USA Cal Club: Round 2  (Read 148050 times)

0 Members and 1 Guest are viewing this topic.

Offline orin

  • Frequent Contributor
  • **
  • Posts: 445
  • Country: us
Re: USA Cal Club: Round 2
« Reply #100 on: June 14, 2018, 04:46:23 am »
I know I said I would follow a random participant order, but I'm going to violate that rule and send it off to kj7e next, to take advantage of his recent calibration  ;D :-DMM

Well, that sounds like a good exception to the rule.


And FWIW, take me out of the rotation for the first two weeks of September.

For obvious reasons, it might be a good idea to let cellularmitosis know of your vacation plans... (privately, even though I just violated that idea).
 
The following users thanked this post: cellularmitosis

Offline vindoline

  • Supporter
  • ****
  • Posts: 328
  • Country: us
Re: USA Cal Club: Round 2
« Reply #101 on: June 14, 2018, 11:42:29 am »
I'm fine with the idea of randomizing the order of the list. It might be a good idea though to pick the order of the list now so that participants have a rough idea of when their turn is. This can help us to be prepared and more efficient during our turn.
 

Online CatalinaWOW

  • Super Contributor
  • ***
  • Posts: 5385
  • Country: us
Re: USA Cal Club: Round 2
« Reply #102 on: June 14, 2018, 03:58:01 pm »
Even if you don't pick or publish the entire list now, it would be good to have a month or two of warning, both for preparation on our end, and to give time for feedback to you if we aren't going to be available.  Obviously all of our plans change from time to time, but most of us do have some major plans of some sort.
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: USA Cal Club: Round 2
« Reply #103 on: June 14, 2018, 04:07:48 pm »
Even if you don't pick or publish the entire list now, it would be good to have a month or two of warning, both for preparation on our end, and to give time for feedback to you if we aren't going to be available.  Obviously all of our plans change from time to time, but most of us do have some major plans of some sort.

That's a good point, but tough considering an unknown turn-around time for each tester.

Perhaps just the list, so one would know they are next?  I'm for example gone about 40% of the warm months, so at least some flexibility needs to be considered.

Offline cellularmitosisTopic starter

  • Supporter
  • ****
  • Posts: 1111
  • Country: us
Re: USA Cal Club: Round 2
« Reply #104 on: June 14, 2018, 04:55:01 pm »
Good ideas / concerns around scheduling -- I'll get the list made and published.
LTZs: KX FX MX CX PX Frank A9 QX
 
The following users thanked this post: bitseeker

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: USA Cal Club: Round 2
« Reply #105 on: June 15, 2018, 04:21:00 am »
But then you will not have surprise guests,  :-DD.
And I'm still waiting on resistors.  :=\
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline cellularmitosisTopic starter

  • Supporter
  • ****
  • Posts: 1111
  • Country: us
Re: USA Cal Club: Round 2
« Reply #106 on: June 15, 2018, 04:32:52 am »
This is a good point :).  There might have to be a few more rule violations
LTZs: KX FX MX CX PX Frank A9 QX
 

Offline bitseeker

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: USA Cal Club: Round 2
« Reply #107 on: June 15, 2018, 05:06:10 am »
Include some entries marked "RESERVED" or "? ? ? ? ?" to enable surprises, sudden schedule changes, and future rule violations.
TEA is the way. | TEA Time channel
 

Offline Svgeesus

  • Regular Contributor
  • *
  • Posts: 78
  • Country: us
Re: USA Cal Club: Round 2
« Reply #108 on: June 15, 2018, 09:45:00 pm »
Speaking of participant order, I will be away on foreign travel from the last week of June until the first week of September. Still able to read and post, but no access to my equipment while traveling.
 

Offline ArthurDent

  • Super Contributor
  • ***
  • Posts: 1193
  • Country: us
Re: USA Cal Club: Round 2
« Reply #109 on: June 16, 2018, 01:29:16 am »
A continually updated list would be a good idea because things can change quickly and unexpectedly. Ten days ago I was putting the final touches on my first Dr Frank board, which was looking good, when I decided to clean all the remaining flux off the board before permanently mounting it in the enclosure. After adding a transistor buffer for the 10V output, making a few changes, and selecting the trim resistors, it needed cleaning. Part way through the cleaning I heard a click from my power supply and suddenly realized I hadn’t disconnected the power before mucking with the board and the click I heard was the supply going into current limit. I checked and the 7 volt output from the LTZ1000A was now 12 volts. DAMN!

My second board is looking good and I hope to have the replacement LTZ1000A for the first board in any day now. I’ll be more careful with this one and both boards should be ready by the time my turn comes up.
 

Offline CalMachine

  • Frequent Contributor
  • **
  • Posts: 477
  • Country: us
  • Metrology Nut
Re: USA Cal Club: Round 2
« Reply #110 on: June 16, 2018, 01:40:32 am »
A continually updated list would be a good idea because things can change quickly and unexpectedly. Ten days ago I was putting the final touches on my first Dr Frank board, which was looking good, when I decided to clean all the remaining flux off the board before permanently mounting it in the enclosure. After adding a transistor buffer for the 10V output, making a few changes, and selecting the trim resistors, it needed cleaning. Part way through the cleaning I heard a click from my power supply and suddenly realized I hadn’t disconnected the power before mucking with the board and the click I heard was the supply going into current limit. I checked and the 7 volt output from the LTZ1000A was now 12 volts. DAMN!

My second board is looking good and I hope to have the replacement LTZ1000A for the first board in any day now. I’ll be more careful with this one and both boards should be ready by the time my turn comes up.

Ouch!!  I've done something similar awhile back. I desoldered the output and power wires after a test, while the power was still on lol.  Thankfully the LTZ seemed to have went unscathed. 
All your volts are belong to me
 

Offline hwj-d

  • Frequent Contributor
  • **
  • Posts: 676
  • Country: de
  • save the children - chase the cabal
Re: USA Cal Club: Round 2
« Reply #111 on: June 16, 2018, 08:24:48 am »
I think, it's probably too expensive to send something to germany back and forth (heard something like 60-80$). So, my participation can't be not more like quiet rhetorical, after spending one of my refs to cellular USA Cal Club (I hope, I still get it on the line, next time :phew:). So, please take me out of list.  Good luck to the club  :-+
« Last Edit: June 16, 2018, 12:39:26 pm by hwj-d »
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3281
  • Country: de
Re: USA Cal Club: Round 2
« Reply #112 on: June 16, 2018, 02:44:03 pm »
I think, it's probably too expensive to send something to germany back and forth (heard something like 60-80$).

It mainly depends on the weight. And the service that you use.
from here to US it is 3,70 Eur below 500g (if you send it as letter) + 2,50 Eur for tracking (Einschreiben)
or 7 Eur below 1000g. (+ the tracking).
Of course the value of the content should be below the customs limits otherwise there are customs fees.

From US to D I payed between 7.10$ and 23.50$ for resistors from Edwin depending on USPS service.
So it is somewhat more but not 60$.

with best regards

Andreas


 

Offline cellularmitosisTopic starter

  • Supporter
  • ****
  • Posts: 1111
  • Country: us
Re: USA Cal Club: Round 2
« Reply #113 on: June 16, 2018, 06:13:25 pm »
The other night I threw the ref into a small styrofoam ice chest along with the temperature logger and a tupperware container full of hot water.  A temperature controller would be better, but this was something I could do quickly in the time I had available.  The ref was hooked up to my Keithley 2015 (the same I've used in previous measurements for this ref).

Interestingly, it looks like the ref by itself has a negative temperature coefficient (in this temperature range).  Previously, when I had adjusted the aircon (and thus measured the combined tempco of the ref and my meter), I had measured a positive tempco: https://www.eevblog.com/forum/metrology/usa-cal-club-round-2/msg1516561/#msg1516561

The ref is packaged up and I'm about to run by the post office to send it off to kj7e (I hope they are still open on Saturday past noon!).
LTZs: KX FX MX CX PX Frank A9 QX
 
The following users thanked this post: kj7e

Offline kj7e

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: USA Cal Club: Round 2
« Reply #114 on: June 16, 2018, 07:07:09 pm »
Looking forward to it.  I have a short list of tests I want to run.  0.1-10Hz noise, 1 hour and 24 hour stability and comparison to my references at a steady 23C,  TC sweeps from 10C to 40C and back, another stability test at 23C after the TC sweep, loading the output at 10G, 10M, 1M and 100K effects, maybe add in some EMI tests with shielded and non-shielded test leads.
 
The following users thanked this post: flittle

Offline hwj-d

  • Frequent Contributor
  • **
  • Posts: 676
  • Country: de
  • save the children - chase the cabal
Re: USA Cal Club: Round 2
« Reply #115 on: June 16, 2018, 09:36:59 pm »
I think, it's probably too expensive to send something to germany back and forth (heard something like 60-80$).

It mainly depends on the weight. And the service that you use.
from here to US it is 3,70 Eur below 500g (if you send it as letter) + 2,50 Eur for tracking (Einschreiben)
or 7 Eur below 1000g. (+ the tracking).
Of course the value of the content should be below the customs limits otherwise there are customs fees.

From US to D I payed between 7.10$ and 23.50$ for resistors from Edwin depending on USPS service.
So it is somewhat more but not 60$.

with best regards

Andreas
That sounds better.
My info was from ups. They told me, the cheapest from germany to USA, value of goods 10€, is 'UPS Expedited', plus fuel surcharge 15€, plus private delivery, incl., something about 90€. (Send it as letter with this outer dimension like a packed Hammond 1590B not possible) :wtf:
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3281
  • Country: de
Re: USA Cal Club: Round 2
« Reply #116 on: June 16, 2018, 09:44:26 pm »
A international letter can also be in a card box (not only a jiffy bag)
(maximum size l + w + h = 90 cm)

https://www.deutschepost.de/de/produkte.html

 

Offline hwj-d

  • Frequent Contributor
  • **
  • Posts: 676
  • Country: de
  • save the children - chase the cabal
Re: USA Cal Club: Round 2
« Reply #117 on: June 16, 2018, 09:58:04 pm »
A international letter can also be in a card box (not only a jiffy bag)
(maximum size l + w + h = 90 cm)
Will ask Deutsche Post for that.
But, that also has to be done by customs. As i understood, handling flats for that are the expensive part. I don't want to cellular, that he suddenly has to pay customs and other costs.
 

Offline cellularmitosisTopic starter

  • Supporter
  • ****
  • Posts: 1111
  • Country: us
Re: USA Cal Club: Round 2
« Reply #118 on: June 16, 2018, 11:10:26 pm »
When I shipped resistors to Andreas it wasn’t that expensive, I don’t recall the exact amount but it was deifninitely under $20.  At the rate this club is going, this is a pretty cheap hobby, so don’t worry about it Hwj-d :)
LTZs: KX FX MX CX PX Frank A9 QX
 
The following users thanked this post: hwj-d

Offline vindoline

  • Supporter
  • ****
  • Posts: 328
  • Country: us
Re: USA Cal Club: Round 2
« Reply #119 on: June 17, 2018, 01:10:13 am »
Looking forward to it.  I have a short list of tests I want to run.  0.1-10Hz noise, 1 hour and 24 hour stability and comparison to my references at a steady 23C,  TC sweeps from 10C to 40C and back, another stability test at 23C after the TC sweep, loading the output at 10G, 10M, 1M and 100K effects, maybe add in some EMI tests with shielded and non-shielded test leads.

Sounds like a great set of benchmark tests for the ref!
 

Offline hwj-d

  • Frequent Contributor
  • **
  • Posts: 676
  • Country: de
  • save the children - chase the cabal
Re: USA Cal Club: Round 2
« Reply #120 on: June 17, 2018, 03:14:35 am »
When I shipped resistors to Andreas it wasn’t that expensive, I don’t recall the exact amount but it was deifninitely under $20.  At the rate this club is going, this is a pretty cheap hobby, so don’t worry about it Hwj-d :)
I've been careful with this, after arrow, at this time as the ltz1000 had elsewhere delivery problems, send me one with dhl in a 'huge' packet. Dhl demanded customs duty and handling flat at delivery time, that more as doubles the price. I didn't accept the part with that delivery costs. After that i saw in online delivery protocoll, that dhl "parked" that package for more than one month in their hand-over stock, before they give it back to arrow. Luckily i got my money back from paypal.

Edit: But with ~20$ i'm in again. Thank you cellular.  :scared:
« Last Edit: June 17, 2018, 03:37:43 am by hwj-d »
 

Online 2N3055

  • Super Contributor
  • ***
  • Posts: 7032
  • Country: hr
Re: USA Cal Club: Round 2
« Reply #121 on: June 17, 2018, 08:17:41 am »


Try Mouser next time. It has free shipping over 50USD. I'm in Croatia, so pretty much same as Germany...
They ship some from EU, and some from USA, but no difference for you.  They don't have LTZ1000 on stock now but they will at the end of July...

Regards,

Sinisa
 
The following users thanked this post: hwj-d

Offline hwj-d

  • Frequent Contributor
  • **
  • Posts: 676
  • Country: de
  • save the children - chase the cabal
Re: USA Cal Club: Round 2
« Reply #122 on: June 17, 2018, 09:01:14 am »
Try Mouser next time. It has free shipping over 50USD. I'm in Croatia, so pretty much same as Germany...
They ship some from EU, and some from USA, but no difference for you.  They don't have LTZ1000 on stock now but they will at the end of July...

Regards,

Sinisa
Hi Sinisa,
yes, Digikey the same. They need 3 days to deliver from Thief River Falls to me. Big german distributors haven't completet the package in this time...  :-//
 
The following users thanked this post: 2N3055

Offline kj7e

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: USA Cal Club: Round 2
« Reply #123 on: June 22, 2018, 12:07:23 am »
Package arrived this afternoon.  First test was a power on warmup plot.  The reference had been sitting in the lab at 23C for about an hour, then powered up to see how long it would take before it settled down.  First screen shot is 1 hour, so it took about 30 min for things to warm up in the tin enclosure.  Second screen shot is a 5 hour plot,  my wife turned off the room A/C around the 13K sec mark, you can see how the plot stabilized some.  Tomorrow I will have to see if this is the reference or my meter showing the thermal sensitive.

Off the bat, I am measuring ~5ppm lower than cellularmitosis.

« Last Edit: June 22, 2018, 12:09:04 am by kj7e »
 
The following users thanked this post: Vgkid, bitseeker, hwj-d, flittle

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2724
  • Country: us
Re: USA Cal Club: Round 2
« Reply #124 on: June 22, 2018, 12:39:43 am »
Looks like I don't need to worry about my multimeters being too drifty.
If you own any North Hills Electronics gear, message me. L&N Fan
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf