Author Topic: Ultra Precision Reference LTZ1000  (Read 1400843 times)

0 Members and 5 Guests are viewing this topic.

Offline Grandchuck

  • Frequent Contributor
  • **
  • Posts: 690
  • Country: us
 

Offline Villain

  • Regular Contributor
  • *
  • Posts: 71
  • Country: de
Re: Ultra Precision Reference LTZ1000
« Reply #2876 on: February 20, 2020, 06:14:42 pm »
Oh very interesting, thank you
 

Offline Edwin G. Pettis

  • Frequent Contributor
  • **
  • Posts: 452
  • Country: us
  • The plural of anecdote is not data.
Re: Ultra Precision Reference LTZ1000
« Reply #2877 on: February 20, 2020, 06:26:58 pm »
As expected, the ADI implementation of the LTZ1000 circuits are flawed, they do not conform to the original design and does not produce the original specifications.  I would not recommend buying this board if you really want a good LTZ1000 reference.  They don't know what they're doing yet, they have no experience with this design and should have kept to the original design of LT.  They used the wrong resistors too....I know you were expecting me to say it but it is a fact none the less.
 
The following users thanked this post: hwj-d, MegaVolt

Offline Villain

  • Regular Contributor
  • *
  • Posts: 71
  • Country: de
Re: Ultra Precision Reference LTZ1000
« Reply #2878 on: February 20, 2020, 06:45:40 pm »
Yeah had a quick look at the design. I must say that i am too lazy to build one myself though, its time consuming..

« Last Edit: February 20, 2020, 06:50:14 pm by Villain »
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3281
  • Country: de
Re: Ultra Precision Reference LTZ1000
« Reply #2879 on: February 20, 2020, 07:40:56 pm »
 

Offline splin

  • Frequent Contributor
  • **
  • Posts: 999
  • Country: gb
Re: Ultra Precision Reference LTZ1000
« Reply #2880 on: February 20, 2020, 07:53:37 pm »
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2418
  • Country: de
  • Sounds like noise
Re: Ultra Precision Reference LTZ1000
« Reply #2881 on: February 20, 2020, 07:54:40 pm »
« Last Edit: February 20, 2020, 07:57:42 pm by branadic »
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 
The following users thanked this post: Villain

Offline Villain

  • Regular Contributor
  • *
  • Posts: 71
  • Country: de
Re: Ultra Precision Reference LTZ1000
« Reply #2882 on: February 20, 2020, 07:58:10 pm »
I contacted them already, found it an hour or so ago too.
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3281
  • Country: de
Re: Ultra Precision Reference LTZ1000
« Reply #2883 on: February 20, 2020, 08:06:47 pm »
Another possibility:

http://www.friedrich-messtechnik.de/index.php/unsere-produkte/32-uncategorised/71-esn-2010

With best regards

Andreas

Any idea how much it costs?

I only found it by accident when looking for LTZ1000. So I have no idea.

With best regards

Andreas
 

Offline Villain

  • Regular Contributor
  • *
  • Posts: 71
  • Country: de
Re: Ultra Precision Reference LTZ1000
« Reply #2884 on: February 20, 2020, 08:13:45 pm »
I will let you all know the price when they respond.

edit: IslandLabs guys say 245 eur for LTZ1000ACH board

edit2: the friedrich-messtechnik guy said they only sell to companies
« Last Edit: February 21, 2020, 03:48:12 pm by Villain »
 

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1966
  • Country: us
Re: Ultra Precision Reference LTZ1000
« Reply #2885 on: February 20, 2020, 08:36:10 pm »
Having bought two from island labs I will say they seem to be good quality. When purchased you get data on aging, noise, current calibration, bom, and schematics. They're also fairly helpful with questions and offer 1 free calibration(I forget the time span).

Attached a couple of pictures I took last year of the LTZ1000A board. If I have time I'll take more detailed ones later of both but since they're in two boxes and one is soldered I don't have time just now.
« Last Edit: February 22, 2020, 05:59:32 pm by maginnovision »
 
The following users thanked this post: Andreas, Villain, MiDi, ldmpam

Offline notfaded1

  • Frequent Contributor
  • **
  • Posts: 559
  • Country: us
Re: Ultra Precision Reference LTZ1000
« Reply #2886 on: February 23, 2020, 07:00:47 am »
It just wouldn't be fun not build some of my own... I am interested in converting this into a non-LTZ reference though.  That's why I asked about best way to safely cut PCB's... IDK about tin snips but I know it's been done.  It's in a working 8842A now.  Somehow the reference might be more fun than the old MM to me I think.

Bill
.ılılı..ılılı.
notfaded1
 

Offline notfaded1

  • Frequent Contributor
  • **
  • Posts: 559
  • Country: us
Re: Ultra Precision Reference LTZ1000
« Reply #2887 on: February 23, 2020, 07:07:04 am »
It just wouldn't be fun not build some of my own... I am interested in converting this into a non-LTZ reference though.  That's why I asked about best way to safely cut PCB's... IDK about tin snips but I know it's been done.  It's in a working 8842A now.  Somehow the reference might be more fun than the old MM to me I think.

Bill
Didn't you build an enclosed 10V reference with this method @Micke?
.ılılı..ılılı.
notfaded1
 

Offline Micke

  • Regular Contributor
  • *
  • Posts: 101
  • Country: se
Re: Ultra Precision Reference LTZ1000
« Reply #2888 on: February 23, 2020, 09:30:11 am »
Yepp, I had a 8842A DMM beyond rescue...
I just used a hacksaw to cut out RefAmp, thought of using Dremel as well, but thought that hacksaw would generate less vibrations, and easier make a straight cut.
Could have made RefAmp board even smaller, but wanted to be on the safe side, and to have somewhere to put mounting screws...
I soldered power +/- 15V directly on Op-amp pins U702 on the secondary side. For Ground, I used the via holes just under the text "C103" (think shield was soldered in these holes originally)
I am not very impressed by the stability though, typical run with FLUKE 8846A below (PLC100), SZA263 in metal box with lid on, powered 24/7 for weeks:
* Min  10.00005V
* Max 10.00010V
* Avg 10.00007V
* SD 7,6µV
* 40000 # samples
So a span of 5ppm between Max and Min, did expect better actually!
 
The following users thanked this post: CalMachine, notfaded1

Offline iMo

  • Super Contributor
  • ***
  • Posts: 4989
  • Country: cv
Re: Ultra Precision Reference LTZ1000
« Reply #2889 on: February 23, 2020, 12:02:00 pm »
Quote
So a span of 5ppm between Max and Min, did expect better actually!
SZA263 is a zener with a transistor on the chip. You cannot expect "LTZ1000 stability" unless the module is ovenised, imho.
 

Offline SilverSolder

  • Super Contributor
  • ***
  • Posts: 6126
  • Country: 00
Re: Ultra Precision Reference LTZ1000
« Reply #2890 on: February 23, 2020, 01:19:48 pm »
[...]
* Min  10.00005V
* Max 10.00010V
* Avg 10.00007V
* SD 7,6µV
* 40000 # samples
So a span of 5ppm between Max and Min, did expect better actually!

5ppm over several weeks is not too shabby?  Is the area temperature controlled?
 

Offline Micke

  • Regular Contributor
  • *
  • Posts: 101
  • Country: se
Re: Ultra Precision Reference LTZ1000
« Reply #2891 on: February 23, 2020, 07:04:09 pm »
[...]
* Min  10.00005V
* Max 10.00010V
* Avg 10.00007V
* SD 7,6µV
* 40000 # samples
So a span of 5ppm between Max and Min, did expect better actually!

5ppm over several weeks is not too shabby?  Is the area temperature controlled?

Did not record start/stop time, but 40000 samples at PLC100 is unfortunately only like 22h
Yeah, did not expect LTZ1000 performance, but I am pretty sure the SZA263 drift more than my AD581LH (need to find the notes from the measurenents... ;) )
 

Offline maginnovision

  • Super Contributor
  • ***
  • Posts: 1966
  • Country: us
Re: Ultra Precision Reference LTZ1000
« Reply #2892 on: February 24, 2020, 04:26:39 am »
I've finally started logging voltage and temp together. I'm using 1 array of 10 DS18B20's to measure ambient, 1 array of 5 for each of 2 meters. The placement for the arrays for the meters is essentially placed between the ADC and the reference. Is that the best place to measure the internal temp, or should it be more central to the analog board?
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2418
  • Country: de
  • Sounds like noise
Re: Ultra Precision Reference LTZ1000
« Reply #2893 on: February 28, 2020, 06:18:59 pm »
Hi,

I tested Analog Discovery 2 spectrum analyzer function together with pipelies LNA (0.1 ... 10Hz) and found that this is a decent setup, without the need of a boat anchor on the desk. Thought that could be of interest for you guys. As an example I measured LNA with shorted input, a 9V battery and one of my LTZ references.

-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 
The following users thanked this post: 2N3055, notfaded1

Offline SilverSolder

  • Super Contributor
  • ***
  • Posts: 6126
  • Country: 00
Re: Ultra Precision Reference LTZ1000
« Reply #2894 on: February 28, 2020, 07:06:41 pm »

[...]  together with pipelies LNA (0.1 ... 10Hz)  [...]


What is the "pipelies LNA" ?
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2418
  • Country: de
  • Sounds like noise
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 
The following users thanked this post: SilverSolder

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2418
  • Country: de
  • Sounds like noise
Re: Ultra Precision Reference LTZ1000
« Reply #2896 on: April 04, 2020, 12:15:56 pm »
I always wondered how the noise spectrum of my LTZ references would look like. Sure we can use DMM readings to get a raw idea, but even the 1/f region is always hard to get. Now that I have one of those nice Digital Spectrum Analyzers (Advantest R9211) or as they are also called Dynamic Signal Analyzers or FFT Analyzers, I was acutally able to create such a nice noise plot today.

-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 
The following users thanked this post: Dr. Frank, Andreas, e61_phil, SilverSolder, razvan784, MiDi, MegaVolt, iMo

Offline SilverSolder

  • Super Contributor
  • ***
  • Posts: 6126
  • Country: 00
Re: Ultra Precision Reference LTZ1000
« Reply #2897 on: April 04, 2020, 12:26:39 pm »
I always wondered how the noise spectrum of my LTZ references would look like. Sure we can use DMM readings to get a raw idea, but even the 1/f region is always hard to get. Now that I have one of those nice Digital Spectrum Analyzers (Advantest R9211) or as they are also called Dynamic Signal Analyzers or FFT Analyzers, I was acutally able to create such a nice noise plot today.

-branadic-

That is very cool!  -  what is the noise floor of an instrument like that?

 

Offline splin

  • Frequent Contributor
  • **
  • Posts: 999
  • Country: gb
Re: Ultra Precision Reference LTZ1000
« Reply #2898 on: April 04, 2020, 01:34:56 pm »
Hi,

I tested Analog Discovery 2 spectrum analyzer function together with pipelies LNA (0.1 ... 10Hz) and found that this is a decent setup, without the need of a boat anchor on the desk. Thought that could be of interest for you guys. As an example I measured LNA with shorted input, a 9V battery and one of my LTZ references.

-branadic-

An LTZ1000 with no flicker noise?   :-//
 

Online Andreas

  • Super Contributor
  • ***
  • Posts: 3281
  • Country: de
Re: Ultra Precision Reference LTZ1000
« Reply #2899 on: April 04, 2020, 01:36:20 pm »
Mhm,

I can explain the 50 Hz peak in the spectrum: your cookies box isnt tight enough.
But where do the 60 Hz come from? (Screen frequency of the instrument?)

with best regards

Andreas
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf