Author Topic: Repair worklog : Old rusty HP 3458A  (Read 250959 times)

0 Members and 1 Guest are viewing this topic.

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5504
  • Country: de
Re: Repair : Old rusty HP 3458A
« Reply #550 on: October 19, 2016, 08:25:57 am »
No and no. You guys such savages  :-/O.

One local taiwanese volt-nut contacted me few days ago. After some challenging discussion with great help of google translate, it was determined that he has some nice gear, including a RAM-sick 3458A. So hopefully next month we will have a chance to meet and look on his 3458A, and perhaps get my 4920M calibrated. This is my best shot right now for AC calibration.  :'(
How does fixing his 3458A help you to get your 4920M calibrated?
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair : Old rusty HP 3458A
« Reply #551 on: October 19, 2016, 10:12:05 am »
HighVoltage
That is to be determined. It's all I can say right now, sorry.  :bullshit:
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline klaus11

  • Supporter
  • ****
  • Posts: 156
  • Country: 00
Re: Repair : Old rusty HP 3458A
« Reply #552 on: October 19, 2016, 11:19:46 am »
Another question, the Front / Rear button when pressed remains locked, or else you press and exits to stop pushing.
I need to know if so or is defective.
HP3458A, HP3245a, Keithley 2000, Fluke 87V, Rigol DP832, TEK TDS5052B, HP33120A
 

Offline Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2402
  • Country: de
Re: Repair : Old rusty HP 3458A
« Reply #553 on: October 19, 2016, 11:45:21 am »
Another question, the Front / Rear button when pressed remains locked, or else you press and exits to stop pushing.
I need to know if so or is defective.

That is a purely mechanical switch, therefore stays locked in either position.
If it flips out again, the locking spring/mechanism is defect.
 

Offline klaus11

  • Supporter
  • ****
  • Posts: 156
  • Country: 00
Re: Repair : Old rusty HP 3458A
« Reply #554 on: October 19, 2016, 11:49:55 am »
Thank DR. Frank, the conclusion is faulty.
HP3458A, HP3245a, Keithley 2000, Fluke 87V, Rigol DP832, TEK TDS5052B, HP33120A
 

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair : Old rusty HP 3458A
« Reply #555 on: October 26, 2016, 12:14:43 pm »
I have a question to 3458A owners. Anyone was using non-10V/non-10K standards for calibration before?
Meter actually allow 1V - 12V for DCV and 1K to 12K for OHM, but also accept -0.9 to 0.9 values and -12 to -1V, which makes me thinking :)

If I understand correctly, less than 10V/10K would compromise adjustment, as measurement value is more suspected by noise floor limit, reducing SNR and absolute accuracy precision. That's clear more or less.

But how about -10V calibration? Anyone tried? :)
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline Macbeth

  • Super Contributor
  • ***
  • Posts: 2571
  • Country: gb
Re: Repair : Old rusty HP 3458A
« Reply #556 on: October 27, 2016, 02:52:58 am »
If I understand correctly, less than 10V/10K would compromise adjustment, as measurement value is more suspected by noise floor limit, reducing SNR and absolute accuracy precision. That's clear more or less.

But how about -10V calibration? Anyone tried? :)

You can always take a backup of your SRAM and then try your experiment surely?  ;)

I have to admit when adjusting the cal of my Keithley 2015 I have merely swapped the +/- leads around when doing the +10V and -10V DC cal. I'm sure I should be shot down as a heretic for this even though I allowed a few minutes for the lead swap to settle any thermals?
 

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair : Old rusty HP 3458A
« Reply #557 on: October 27, 2016, 04:09:47 am »
Quote
You can always take a backup of your SRAM and then try your experiment surely?

Aha, A5 CPU board is on it's way already.  :popcorn:

Quote
I'm sure I should be shot down as a heretic for this even though I allowed a few minutes for the lead swap to settle any thermals?

Why? That's what Fluke 5700A does as well, just swaps the signal with relay.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline Macbeth

  • Super Contributor
  • ***
  • Posts: 2571
  • Country: gb
Re: Repair : Old rusty HP 3458A
« Reply #558 on: October 27, 2016, 07:27:43 pm »
Why? That's what Fluke 5700A does as well, just swaps the signal with relay.
Ahh. I figured something more sophisticated like an actual dual rail supply +/- and using something like "lead compensation" too.

With the 3458A I was always under the impression that you only  ;) needed an an accurate +10V DC and a 10K ohm resistor for complete calibration, no need for a negative calibration at all. But from what you are saying any accurate source of a reasonable magnitude can be used as long as you tell the meter what it is measuring? Like with the Keithleys IIRC anything from 9.9V - 10.1V are acceptable. It seems the 3458A has much more leeway if it allows negative volts to be used as a calibration reference!
 

Offline PA0PBZ

  • Super Contributor
  • ***
  • Posts: 5163
  • Country: nl
Re: Repair : Old rusty HP 3458A
« Reply #559 on: October 27, 2016, 08:08:31 pm »
Interesting, my Datron 1081 insists on having both + and - on all the DC calibration steps.
Keyboard error: Press F1 to continue.
 

Offline View[+]Finder

  • Supporter
  • ****
  • Posts: 186
  • Country: us
    • Sparks! A Learning Place for Curious Minds
Re: Repair : Old rusty HP 3458A
« Reply #560 on: December 27, 2017, 03:25:03 am »
Re: reversing the filter HP3458A, HP3245A
1) Fans prefer having their 'low pressure' side in open air (vacuum is limited, but pressure isn't)
2) Filters work better when they have a little lint on them (the designers knew that)
3) Filter replacement on air entry is difficult if inside the case (someone has to do it)
4) the 'dirt' in the air is mostly human skin cells (get a HEPA filter for your lab)

Just some observations . . .
 

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair : Old rusty HP 3458A
« Reply #561 on: July 06, 2018, 05:02:24 am »
Since so many questions about 3458A lately popping on forums, I think it's justified to revive this worklog thread.

Another reason is that I just finished initial calibration test, using yesterday calibrated 5700A (to standards which values I know sub-ppm (maybe  :popcorn:) ).

Calibration points performance report - non-traceable, using my calkit software.

Please keep in mind that verification specification is 24 hour, 95% with K=1.
It is NOT what usual accredited calibration lab data reports have.  ;)

This meter was DIY calibrated in January 2017. It was 24/7/365 since.
I'm happy with stability, no adjustment calibration is necessary. She is stable!  :box:


« Last Edit: July 06, 2018, 05:05:20 am by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: edavid, Zucca, klaus11

Offline Zucca

  • Supporter
  • ****
  • Posts: 4423
  • Country: it
  • EE meid in Itali
Re: Repair : Old rusty HP 3458A
« Reply #562 on: July 06, 2018, 06:44:55 am »
Quote
95% with K=1

I don't know what to google, what does it means?  :-[ (Imbarassed)
Can't know what you don't love. St. Augustine
Can't love what you don't know. Zucca
 

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair : Old rusty HP 3458A
« Reply #563 on: July 06, 2018, 09:09:14 am »
Confidence level is optimistic 95% (most of specs are confidence level 99%). K number is guardband ratio. I use own guardband , so this multiplier is not used in test reports.

  Most real calibration certificates are at K=2.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5504
  • Country: de
Re: Repair : Old rusty HP 3458A
« Reply #564 on: July 06, 2018, 09:15:51 am »
Quote
95% with K=1

I don't know what to google, what does it means?  :-[ (Imbarassed)
Read the beginners guide to uncertainties...
There are 3 kinds of people in this world, those who can count and those who can not.
 
The following users thanked this post: TiN, Zucca

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair worklog : Old rusty HP 3458A
« Reply #565 on: July 07, 2018, 12:29:29 pm »
Since it's time to recalibrate my newer 3458A it is time to take some photos of the process.
This unit had long overdue needed replacement for front binding posts, so it will be received new replacement today (expensive connectors, ouch)



It's pity that to replace front and rear terminal blocks you pretty much have to take apart whole inguard section, as there is no other way to gain access. Photo below also have rare sight of modified 3458A A9 reference for lower temperature and better annual drift.



And today living room looks like a little mess  :-DD.



Ignore the drill on background, it's there just for intimidation.
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: ManateeMafia, VK5RC, Zucca, CalMachine

Offline martinr33

  • Frequent Contributor
  • **
  • Posts: 363
  • Country: us
Re: Repair worklog : Old rusty HP 3458A
« Reply #566 on: July 08, 2018, 04:11:43 am »
 I don't see the replacement keypad buttons in the new color...

(Keysight has them, but you have  to call as they are not on  the parts list - they only list the old color).
 

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair worklog : Old rusty HP 3458A
« Reply #567 on: July 08, 2018, 06:37:07 am »
I realized this yesterday, but it's no big deal :) Only trained volt-nut eye can pick a difference  :-DD
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline VK5RC

  • Supporter
  • ****
  • Posts: 2672
  • Country: au
Re: Repair worklog : Old rusty HP 3458A
« Reply #568 on: July 08, 2018, 10:52:30 am »
Thank for the close up of the A9 reference - Just noted the 111 \$\Omega\$ resistor -I will need to look at other 03458-66509 - I have always assumed they all had 120  \$\Omega\$
Whoah! Watch where that landed we might need it later.
 

Offline Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2402
  • Country: de
Re: Repair worklog : Old rusty HP 3458A
« Reply #569 on: July 08, 2018, 11:48:57 am »
Sure, the 3458A reference has always 111 Ohm, as its temperature is 95°C, therefore the Ube is about 70mV lower than intended , so to maintain the nominal zener current, the resistor has to be lower than the nominal 120 Ohm, also.
By pimping the standard ref to 65°C (100k || 15k), this increases the zener current and decreases noise.
Maybe that's the reason, why my instrument seems to perform better than others in in the Noise Performance Testing, we've done a few years back.

Greetings from Madeira
Frank
 
The following users thanked this post: TiN, VK5RC, Zucca

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Repair worklog : Old rusty HP 3458A
« Reply #570 on: July 18, 2018, 11:26:39 am »
I got my rubber-bumper for 3458A. Looks kinda strange, what you guys think?

YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: Zucca

Offline Zucca

  • Supporter
  • ****
  • Posts: 4423
  • Country: it
  • EE meid in Itali
Re: Repair worklog : Old rusty HP 3458A
« Reply #571 on: July 18, 2018, 11:53:03 am »
Father:
Quote
There is a room in Taiwan full of expensive big and small toys. Hell expensive toys, with a lot of numbers and colored small lights creating an unique atmosphere.
In this room a russian man is always improving and fixing those toys.
Sometime he writes his incredible stories about chasing that failure or that strange ppm jump.
Few people in this world can understand him and his complicated games, but a lot of people just admire him for his passion and talent.
One day he even protected one of his favorite 8,5 DMM with a long and soft rubber-bumper.

The almost sleeping son looks at his dad with big eyes.

Father:
Quote
Yeah, it's too strange to be true. You are right.
Good night son, I tell you the rest of the story tomorrow.

I smoked too much, don't ask me what
Can't know what you don't love. St. Augustine
Can't love what you don't know. Zucca
 

Offline hwj-d

  • Frequent Contributor
  • **
  • Posts: 676
  • Country: de
  • save the children - chase the cabal
Re: Repair worklog : Old rusty HP 3458A
« Reply #572 on: July 18, 2018, 12:41:34 pm »
Father:
Quote
There is a room in Taiwan full of expensive big and small toys. Hell expensive toys, with a lot of numbers and colored small lights creating an unique atmosphere.
In this room a russian man is always improving and fixing those toys.
Sometime he writes his incredible stories about chasing that failure or that strange ppm jump.
Few people in this world can understand him and his complicated games, but a lot of people just admire him for his passion and talent.
One day he even protected one of his favorite 8,5 DMM with a long and soft rubber-bumper.

The almost sleeping son looks at his dad with big eyes.

Father:
Quote
Yeah, it's too strange to be true. You are right.
Good night son, I tell you the rest of the story tomorrow.

I smoked too much, don't ask me what

 ;D ;D ;D ;D
 

Offline Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2402
  • Country: de
Re: Repair worklog : Old rusty HP 3458A
« Reply #573 on: July 18, 2018, 04:59:56 pm »
I got my rubber-bumper for 3458A. Looks kinda strange, what you guys think?



Looks like a stretched 34401A..

On mine I've got the handles, but your bumper handles with care ..  :-DD

Frank
 

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3763
  • Country: ca
  • Living the Dream
Re: Repair worklog : Old rusty HP 3458A
« Reply #574 on: July 18, 2018, 05:46:49 pm »
I don't think TiN really like that boot. I think he just put it on there to tease me. I have boots on nearly all my gear - even a Keithley 2400 has one. My 3458A's have rack handles though. With the boot on the 3458A it doesn't look like a serious piece of gear anymore.
VE7FM
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf