Mickle T., the PRI5610 could be a version of the BK7, yes, but both have different package with different pin quantity and we don't know for sure as there is no datasheet for the BK7. The patents don't show more details either.
I guess you mix up a few points Kleinstein.
1. The latest firmware I flashed into the OTP has the original factory cal constants copied to it that I restored from the original EPROM before.
2. There could be a small offset at 0 V coming from the source (Analogics AN3200) that has to be ignored and is not the problem as I compare the Prema 6048 against the Solatron 7081, not against the set value at the source.
3. When I short the input the Prema 6048 reads zero, so again, don't worry about the offset.
4. The large INL at 1 V, falling of towards 10 V, is real and not just a coincidence, but I'm performing a smaller step INL measurement right about now, will see how that turns out. I also increased soak time and take a few more samples at each voltage.
-brandic-