Thought everyone might enjoy some details and pictures of the the ADCMT 7480T 8.5 digit multimeter.

ADCMT separated from Advantest in 2003.

The 7480T was made by ADCMT for Advantest, they don't sell it and I've been told to contact Advantest for information. I have not received any reply from Advantest. There are no public docs on the meter, I have only found details on its little brother the 7470 7.5 digit version.

The details I do know:

LTZ1000CH ref  - 13K/1K temp set point.
9K ohm hermetic AE branded resistor ref.
Artifact calibration - 9-11 VDC/9-11Kohms.
DCV and 2/4 wire ohms support. 1000 VDC/10 meg ohm top ranges.
Has internal calibration which is recommended to be used every degree C/ 24 hours. A full DCV/resistance internal cal takes 4.5 minutes.
When the meter boots it says 6581T on the display, it also reports 6581T via GPIB. USB is disabled.
Fan airflow only runs over the power supply/digital section.
Front/rear jacks selected electronically(rather useful).

No idea when it was calibrated but it arrived reading within 1 PPM of the last cal value of my 732A - hard to complain about that.


« Last Edit: January 04, 2018, 06:11:24 am by TheSteve »
VE7FM
 
The following users thanked this post: quarks, SeanB, Echo88, Mickle T., daqq, 3roomlab, edavid, doktor pyta, kj7e, denimdragon, szszjdb

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #1 on: January 04, 2018, 05:15:23 am »
pics:

VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #2 on: January 04, 2018, 05:16:50 am »
pics
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #3 on: January 04, 2018, 05:17:51 am »
pics
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #4 on: January 04, 2018, 05:19:13 am »
pics
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #5 on: January 04, 2018, 05:20:25 am »
pics
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #6 on: January 04, 2018, 05:22:06 am »
pics
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #7 on: January 04, 2018, 05:23:35 am »
pics
« Last Edit: January 04, 2018, 05:29:36 am by TheSteve »
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #8 on: January 04, 2018, 05:24:53 am »
pics
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #9 on: January 04, 2018, 05:26:43 am »
pics

And docs - I had google translate the 7470(not the 7480) manual which is 98% the same in term of the menus and such. I can find no actual 7480 documents.
« Last Edit: January 04, 2018, 07:17:01 am by TheSteve »
VE7FM
 
The following users thanked this post: quarks, 3roomlab, kj7e, capt bullshot

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2726
  • Country: us
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #10 on: January 04, 2018, 05:44:34 am »
Thanks for all of those pics , the board in img 2426 with the transistor array in interesting.
« Last Edit: January 04, 2018, 06:46:38 am by Vgkid »
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline Pipelie

  • Regular Contributor
  • *
  • Posts: 172
  • Country: cn
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #11 on: January 04, 2018, 06:07:21 am »
Thanks for all of those pics. :-+ I saw a ADCMT 7408T a yeas ago on 38Hot , here it is: http://bbs.38hot.net/thread-193743-1-1.html
The seller asking 2.88W RMB for this unit, last time I checked it still there. I have a change to have a look  what's inside recently, it appears  to be a Simplified version of 6581?
 

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #12 on: January 04, 2018, 06:10:26 am »
Thanks for all of those pics. :-+ I saw a ADCMT 7408T a yeas ago on 38Hot , here it is: http://bbs.38hot.net/thread-193743-1-1.html
The seller asking 2.88W RMB for this unit, last time I checked it still there. I have a change to have a look  what's inside recently, it appears  to be a Simplified version of 6581?

I believe an updated version of the 6581 is exactly what it is. I have a feeling Advantest has a need for it, possibly for internal testing etc and commissioned ADCMT to make it for them.
It would be really nice to find published specs for the unit.
VE7FM
 

Offline zhtoor

  • Frequent Contributor
  • **
  • !
  • Posts: 337
  • Country: pk
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #13 on: January 04, 2018, 06:58:09 am »
LTZ1000CH rebranded PHILIPS  :wtf:

anybody know of such an arrangement?

-zia
 

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #14 on: January 04, 2018, 07:00:21 am »
LTZ1000CH rebranded PHILIPS  :wtf:

anybody know of such an arrangement?

-zia

Umm, try again. The ref was made in the Philippines.
VE7FM
 

Offline Pipelie

  • Regular Contributor
  • *
  • Posts: 172
  • Country: cn
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #15 on: January 04, 2018, 07:01:09 am »
Thanks for all of those pics. :-+ I saw a ADCMT 7408T a yeas ago on 38Hot , here it is: http://bbs.38hot.net/thread-193743-1-1.html
The seller asking 2.88W RMB for this unit, last time I checked it still there. I have a change to have a look  what's inside recently, it appears  to be a Simplified version of 6581?

I believe an updated version of the 6581 is exactly what it is. I have a feeling Advantest has a need for it, possibly for internal testing etc and commissioned ADCMT to make it for them.
It would be really nice to find published specs for the unit.

 SPEC in the manual you uploaded, check page 80. hmm, that's the manual of 7470 :palm:
http://www.adcmt.com/techinfo/product/catalog_ducument/pdf/catalog_Gene_7470.pdf
« Last Edit: January 04, 2018, 07:07:26 am by Pipelie »
 

Offline 3roomlab

  • Frequent Contributor
  • **
  • Posts: 845
  • Country: 00
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #16 on: January 04, 2018, 07:02:44 am »
hmmm the manual only states accuracy spec to 7.5 digits. did i miss anything? or it wasnt intended for 8.5D operation? (i assume 8.5D in GPIB?)
and running my eyes over the R7/R4 in pic 2442, is that really a precision 120 ohm in SMD format?
« Last Edit: January 04, 2018, 07:09:31 am by 3roomlab »
 

Offline zhtoor

  • Frequent Contributor
  • **
  • !
  • Posts: 337
  • Country: pk
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #17 on: January 04, 2018, 07:05:35 am »
LTZ1000CH rebranded PHILIPS  :wtf:

anybody know of such an arrangement?

-zia

Umm, try again. The ref was made in the Philippines.

my bad!!! :palm:
 

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3769
  • Country: ca
  • Living the Dream
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #18 on: January 04, 2018, 07:15:56 am »
hmmm the manual only states accuracy spec to 7.5 digits. did i miss anything? or it wasnt intended for 8.5D operation? (i assume 8.5D in GPIB?)
and running my eyes over the R7/R4 in pic 2442, is that really a precision 120 ohm in SMD format?

Those are the specs for the 7470, not the 7480. The 7470 is limited to 7.5 digits, the 7840 allows 8.5 digits.
That is a 120 ohm SMT resistor, right above the super tiny 13k/1k divider resistor.
VE7FM
 

Offline Mickle T.

  • Frequent Contributor
  • **
  • Posts: 478
  • Country: ru
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #19 on: January 04, 2018, 07:48:08 am »
Thanks for all of those pics!
Only one question: what is a lower limit of the integration time in this meter (how fast is it)?
 

Offline Andreas

  • Super Contributor
  • ***
  • Posts: 3284
  • Country: de
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #20 on: January 04, 2018, 06:29:08 pm »
Hello,

interesting.
Are the blue resistors FLCY from Alpha Electronics?

With best regards

Andreas
 

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2421
  • Country: de
  • Sounds like noise
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #21 on: January 04, 2018, 06:48:19 pm »
Quote
Are the blue resistors FLCY from Alpha Electronics?

They are networks with 2 resistors, so maybe SLD series? You can see the designations "7K34 XF 10K0", "17K0 XF 19K0" and "..K0 XF 17K0".  The last one is not fully visible on none of the pictures.



I also wonder what this magic board with SOT23-3 ic is for.

-branadic-
« Last Edit: January 04, 2018, 06:51:00 pm by branadic »
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
 

Offline Mickle T.

  • Frequent Contributor
  • **
  • Posts: 478
  • Country: ru
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #22 on: January 04, 2018, 06:48:46 pm »
Its SLD and SS resistor networks from Alpha Electronics.
I think "..K0 XF 17K0" - is a 10k-17k.
« Last Edit: January 04, 2018, 06:53:16 pm by Mickle T. »
 
The following users thanked this post: quarks, TiN, Andreas

Offline Andreas

  • Super Contributor
  • ***
  • Posts: 3284
  • Country: de
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #23 on: January 04, 2018, 07:36:42 pm »
I also wonder what this magic board with SOT23-3 ic is for.

Perhaps a multiplexer built with discrete FETs?

with best regards

Andreas
 
The following users thanked this post: TiN

Offline branadic

  • Super Contributor
  • ***
  • Posts: 2421
  • Country: de
  • Sounds like noise
Re: ADCMT 7480T 8.5 digit multimeter
« Reply #24 on: January 04, 2018, 08:13:19 pm »
Quote
I think "..K0 XF 17K0" - is a 10k-17k.

Okay... you seem to have a schematic of this section, can you link it here?