So I built myself a pair of kelvin clips, using the widely available cheap type of clips.
eBay auction: #http://www.ebay.com/itm/2-PCS-IC-SMD-Chip-Alligator-LCR-Kelvin-Clip-Copper-Gold-Plated-Clip-Test-Clamp-/322147974085?hash=item4b0183fbc5:g:PKAAAOSwnFZXXWbCWhile they are decent, They suffer from a lot of axial movement , which causes them to fall off of the terminals. Unfortunately that lead to a bunch of widely varying readings. That was remedied by using the adjacent contact lug as a brace( If you guys want I can illustrate this).
The silicone wire I used was 14gauge wire, I loved this. I had never actually worked with silicone wire before, it is rather enjoyable. The one downside to the many strands of wire that is used, is that you are very likely to lose a few pieces when stripping off the insulation(I use a mechanical stripper, those with hot wire strippers should be more fortunate).
![](http://[url=http://i.ebayimg.com/images/g/R9wAAOSwdzVXiUXi/s-l300.jpg]http://i.ebayimg.com/images/g/R9wAAOSwdzVXiUXi/s-l300.jpg[/url])
The 4mm banana terminals I used are the AMASS type. I tried them on all of my multimeters, and power supplies. They were a tight fight on all, except for a 50yo precision lambda power supply. I plan on buying more of the banana jacks, I figured that they will easily get a bunch of use.
Now for the testing.
I used the Ballantine 3205B set in pulse mode for the resistance measurements, as If I used my 3456A, it would be down in the noise.
So lets get on with the testing…
![](http://i.imgur.com/sO0dwf6.png)
We can see the switch contact resistances, broken up into positions and poles. As expected the 0-position is constantly at the lowest switch resistances. As I was testing I realized that the resistances were rather uniform. With 84% of measurements falling between 0.70milliOhms, and 0.85milliOhms.
The one measurement that always seemed to stand out was that the Pole(3) contact resistances were always higher than the rest. I figured it would either be 2 or 3 due to the distance between bearing surfaces on the central shaft. Though thinking about this it makes since, as the axle of the switch is not as precise, when compared to the other type 31 switches I have disassembled. That will be later, I will not do a complete teardown(due to the much different nature of this type.
![](http://i.imgur.com/EZwkEKG.png)
Lastly I will present the data in a histogram, that is not broken down at all. For a switch that is specified as 1 +/-.50 mOhms, it easily meets those specs.
Next update will be a short one, just going to do a noninvasive internal comparision.