Author Topic: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...  (Read 155041 times)

0 Members and 2 Guests are viewing this topic.

Offline Blogs14

  • Newbie
  • Posts: 9
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #100 on: January 23, 2013, 11:33:10 am »
Well that's embarrassing, don't know how I missed that file when I was looking the first time.

Anyway here's the link and checksum for anyone else:
http://www.mikrocontroller.net/attachment/147456/DS1000EUpdate.zip
MD5: eac4f449e60e2fe3cec561cbfa543bed

(From: http://www.mikrocontroller.net/topic/261138)

Problem is I still can't get the trigger delay calibration to save between power-ups. It doesn't seem to make a difference which firmware I try.

I know this isn't very serious, but it annoys me that the scope isn't perfect.

Any ideas?

EDIT:
I've managed to sort it out now. In the end I tried FW 4.00 again and set the trigger delay calibration. Then, without powering off, I ran auto calibration and now the settings have been saved. Although the auto calibration changed them slightly, they're pretty much perfect.
« Last Edit: January 23, 2013, 06:46:18 pm by Blogs14 »
 

Offline pomonabill221

  • Frequent Contributor
  • **
  • Posts: 252
  • Country: us
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #101 on: March 17, 2013, 02:28:53 am »
we have to find out the hardware version of your unit (from the serial number or better from PCB version) and then we can fix it.

So, we did this, and it fixed the problem! Restoring the "HardVersion" to 07 (my board was marked "DEM07") allowed me to upgrade to 2.06 without the ADC offset problem that was present before when the field was blank. It seems my assumptions about the software assuming a newer hardware version were correct. My realtrigdelay/equtrigdelay were also wiped to 0 and I decided to also calibrate those -- not sure what I did originally. Thanks, drieg!
I, too have DEM07.
I have performed the hack to a DS1102E and it worked, and applied FW version 2.04.00.03 but do not have a hardware version, just jumbled characters.
When I did the serial update (I know, I know), I did accidently hit ENTER rather than ALT 000, but it seemed to take ok.
Now I am wondering if my cal data is corrupt because of the missing HW version?
How did you do it?  I see all these posts saying they had success, but NO ONE has said HOW it was done, or is this being kept a guarded method so it won't be done for fun?
 

Offline jamess

  • Newbie
  • Posts: 2
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #102 on: January 22, 2014, 01:21:07 am »
I've managed to brick my 1052e attempting a firmware downgrade :palm:


Drieg, you you have a moment I've sent you a PM with the details, any help/advice is appreciated.

--J
 

Offline jamess

  • Newbie
  • Posts: 2
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #103 on: February 01, 2014, 02:59:04 pm »
My bricked DS1052e is all fixed up, huge thanks to drieg.

 

Offline pomonabill221

  • Frequent Contributor
  • **
  • Posts: 252
  • Country: us
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #104 on: February 01, 2014, 05:42:18 pm »
That's not fair!!! leaving us in the dark as to what you did to revive your scope!
Did it involve using the JTAG connection?

Good to hear you got your baby back!
 

Offline driegTopic starter

  • Regular Contributor
  • *
  • Posts: 85
  • Country: cz
    • Silcon Electronics
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #105 on: February 02, 2014, 10:12:26 am »
My bricked DS1052e is all fixed up, huge thanks to drieg.
That's not fair!!! leaving us in the dark as to what you did to revive your scope!
Did it involve using the JTAG connection?

Good to hear you got your baby back!
Yes, if the unit doesn't start at all, you need to reprogramm the flash either in external programmer or via JTAG (as we did in this case).

I'll update my first post with more details when I find the time...
« Last Edit: February 02, 2014, 10:13:59 am by drieg »
Bricked Rigol? This thread might be of any help.
 

Offline pomonabill221

  • Frequent Contributor
  • **
  • Posts: 252
  • Country: us
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #106 on: February 02, 2014, 08:22:08 pm »
Thanks for the info... kinda thought so but wanted to make sure.
 

Offline szist

  • Newbie
  • Posts: 2
  • Country: pl
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #107 on: March 16, 2014, 03:32:13 pm »
I got the Rigol DS1052E after uploading the incorrect firmware . When you turn on the Oscilloscope , nothing is displayed only bright screen. USB does not work. Please help in the recovering the operation via JTAG or reprogramming the flash of chip.
 

Offline szist

  • Newbie
  • Posts: 2
  • Country: pl
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #108 on: April 03, 2014, 04:20:52 pm »
I wanted to thank Drieg in repairing DS1052E. He sent repaired firmware, and I could upload it via JTAG and my oscilloscope alive. Direg is great and I am very grateful to him for taking the time.  : :-+
 

Offline Alliyou David

  • Newbie
  • Posts: 1
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #109 on: November 10, 2014, 03:32:09 am »
Dear Drieg: Can you send the firmware mentioned in your blog to my E_Mail:David760190@hotmail.com?Thank you very much.
 
The following users thanked this post: ivi_yak, Dimas77

Offline merseyless

  • Contributor
  • Posts: 11
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #110 on: November 19, 2014, 12:53:48 am »
Hi there, I have a ds1052e that has been bricked. What steps do I have to go through to fix it?
 
The following users thanked this post: Dimas77

Offline Zaq

  • Newbie
  • Posts: 4
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #111 on: June 12, 2015, 05:57:25 pm »
Can some one help me out to get my Rigol DS1052E working again?
Did the 100MHz hack, that worked fine.
Than i want to upgrade firmware and the scope is compleet
dead.
Only white screen, nothing else.
Can you give me instructions what to do.
I dont have a jtag op external programmer.

With regards,
 

Offline ENB

  • Newbie
  • Posts: 1
  • Country: es
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #112 on: March 01, 2016, 11:27:06 pm »
Hi,

Any news for unbrick our scopes? :(

Thanks in advance
 

Offline pantelei4

  • Regular Contributor
  • *
  • Posts: 93
  • Country: ua
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #113 on: January 22, 2017, 07:07:53 am »
Hi, who knows how to communicate with drieg ?
I broke my Rigol 1052E, corrupted calibration data. |O
Last time he was here November, 2016.
I left him a message, unanswered ...
 

Offline pantelei4

  • Regular Contributor
  • *
  • Posts: 93
  • Country: ua
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #114 on: January 24, 2017, 10:01:40 am »
Thank you, who responded. I restored my oscilloscope itself. After an unsuccessful upgrade was lost HW version=? Therefore, it is uncalibrated. The board has DE03 inscription means HW version=03. With the help of the serial interface and the command :INFO:HARDVERSION 03 value was restored.
Then I set the FW version 00.02.04.01.00 and spent calibration. Everything works fine, it cost me a few days of nerves. I agree with drieg, need to be very careful and neat, when you want something else to do.
« Last Edit: January 24, 2017, 10:26:13 am by pantelei4 »
 
The following users thanked this post: ivi_yak

Offline dberdugo

  • Newbie
  • Posts: 1
  • Country: es
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #115 on: May 17, 2017, 08:36:13 pm »
Hello, thanks all for share your experiences and information about our Rigol oscilloscopes.

My case takes about a DS1052D wich I bought less than a year ago, finally I decided to upgrade to DS1102D and got an expensive brick.
This is DS1052D, HW version 58, FW version 00.04.02 SP1, and the serial number is DS1EU175000xxx

Obviously serial commands to change serial and model didn't work.
I tried downgrade directly using "DS1000DUpdate.RGL", the oscilloscope detected pendrive but didn't perform any flashing process.
I tried to modify the header of the FW file to identify the same version as my unit and later, but nothing happens.

Finally I decided to desolder the flash for myself, made a backup with an external programmer, and changes the serial letters (to DS1EA) and model name (DS1102D) at the binary image, but when solder again, the oscilloscope was dead :(
I retried with many images extracted from official FW upgrade files, removing the header and concatenating the 4MB image of this file with the last 4MB of my original backup wich contains serial, model name and suposed calibration data, but the oscilloscope remained dead.

Finally, I got scared and restores exactly the original backup and the oscilloscope breathed again, got back to live, but when enable any channel (analog or logic analyzer) it shows random garbage waveforms, does not matter if I connect the probes to an stable 5V, GND or test waveform pin :(
I performed many times the autocalibration procedure, this makes many tick-tick sounds, complete the progress bars, but when finished, it shows the same behaviour.

I think the factory calibration data is on flash, and I restored the original backup, so it must have this data intact.
Anyone knows how is happening inside my oscilloscope?

Thanks and greetings.
Daniel
 
The following users thanked this post: ivi_yak, Dimas77

Offline ivi_yak

  • Regular Contributor
  • *
  • Posts: 80
  • Country: 00
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #116 on: July 20, 2017, 06:32:11 pm »
hi
I have a brick after firmware updating of DS1302CA |O
Could you post a manual for reflash or advice what to do?
Please post any info about repairing of software bricked RIGOLs
eevblog
 

Offline ivi_yak

  • Regular Contributor
  • *
  • Posts: 80
  • Country: 00
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #117 on: July 20, 2017, 06:46:06 pm »
hi, could you share what king of programmer you use for reflash and it will be great for instructions
thanks!
eevblog
 

Offline Bruzzel

  • Newbie
  • Posts: 2
  • Country: de
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #118 on: January 18, 2019, 11:25:49 pm »
Hello,

I have flash a wrong firmware to my old DS1042C and now it is dead, nothing works. |O
What can I do to fix my bricked scope?
Where can I find information and correct Hex-data for reprogramming the Flash?

The DS1042C board is very similar to the DS1042E Board but has only a S29GL032A flash
 (not a S29GL064 flash) and has no second front USB-Port near the serial EEPROM.

The last firmware was 02.03.16.

Can anybody help me?


 

Offline jarek007

  • Newbie
  • Posts: 1
  • Country: pl
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #119 on: January 19, 2019, 12:22:30 pm »
Hello, When modifying DS1052E probably lost the calibration data - corrupted calibration data. Uploaded firmware 00.04.02 SP1. Interference occurs without a signal.
Please help.
Who knows how to communicate with Drieg
I apologize for the language, but I write from the translator
Jarek
« Last Edit: January 20, 2019, 10:55:00 am by jarek007 »
 

Offline driegTopic starter

  • Regular Contributor
  • *
  • Posts: 85
  • Country: cz
    • Silcon Electronics
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #120 on: February 06, 2019, 10:17:34 am »
Hello Jarek,

drop me a PM, I will look at it.
Bricked Rigol? This thread might be of any help.
 

Offline driegTopic starter

  • Regular Contributor
  • *
  • Posts: 85
  • Country: cz
    • Silcon Electronics
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #121 on: February 06, 2019, 10:20:04 am »
Hello,

I have flash a wrong firmware to my old DS1042C and now it is dead, nothing works. |O
What can I do to fix my bricked scope?
Where can I find information and correct Hex-data for reprogramming the Flash?

The DS1042C board is very similar to the DS1042E Board but has only a S29GL032A flash
 (not a S29GL064 flash) and has no second front USB-Port near the serial EEPROM.

The last firmware was 02.03.16.

Can anybody help me?

Can you read the flash content? You can do it via JTAG, external programmer or send me the board for repair..
Bricked Rigol? This thread might be of any help.
 

Offline driegTopic starter

  • Regular Contributor
  • *
  • Posts: 85
  • Country: cz
    • Silcon Electronics
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #122 on: February 06, 2019, 10:21:36 am »
hi
I have a brick after firmware updating of DS1302CA |O
Could you post a manual for reflash or advice what to do?
Please post any info about repairing of software bricked RIGOLs
Hello,

sorry for late reply. Have you managed to fix your problem? If not, feel free to contact me..
Bricked Rigol? This thread might be of any help.
 
The following users thanked this post: ivi_yak

Offline ivi_yak

  • Regular Contributor
  • *
  • Posts: 80
  • Country: 00
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #123 on: February 06, 2019, 08:14:28 pm »
 :-+
eevblog
 

Offline ivi_yak

  • Regular Contributor
  • *
  • Posts: 80
  • Country: 00
Re: The Dark Side of the Rigol Hack -- Bricked Scope & How to Fix it...
« Reply #124 on: February 23, 2019, 11:24:59 am »
Thanks Mr. Drieg he fix my dump and scope alive now   :-DMM,
I have bricked Rigol ds1302CA after unsuccessfully fw update.
Using XGecu TL866II Plus http://www.xgecu.com/en/ with TSSOP48 adaptor I got a memory dump of S29GL032A90TFIR40.
Drieg fixed my dump at few days and my scope restored now.

eevblog
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf