Author Topic: EEVblog #576 - Advantest R6142 Current Voltage Generator  (Read 55737 times)

0 Members and 1 Guest are viewing this topic.

Offline robrenz

  • Super Contributor
  • ***
  • Posts: 3035
  • Country: us
  • Real Machinist, Wannabe EE
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #25 on: February 04, 2014, 04:09:01 pm »
Thank you!  :-+

Offline electronics man

  • Frequent Contributor
  • **
  • Posts: 686
  • Country: gb
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #26 on: February 04, 2014, 04:10:52 pm »
Tear it down
follow me on twitter @get_your_byte
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 962
  • Country: de
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #27 on: February 04, 2014, 07:41:58 pm »
Tear it down and tell us how a precision analog to digital converter works :)   (especially this calibration sources with 5.5 and more digits)
 

Offline OrangeJacketGuy

  • Contributor
  • Posts: 14
  • Country: us
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #28 on: February 04, 2014, 11:13:19 pm »
*gathers up best Reagan parody voice* Mister Jones, tear down this Voltage Generator!
 

Offline junggwok

  • Contributor
  • Posts: 33
  • Country: au
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #29 on: February 05, 2014, 01:06:10 am »
I made a an offer for a Advantest R6142 over a year ago on eBay, however the seller declined my offer. The unit has been on sale on eBay for over 3 years. It is sold 3 hours after Dave did a video on it. Haha, gotta love his influence.

 

Offline gemby

  • Contributor
  • Posts: 24
  • Country: hr
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #30 on: February 05, 2014, 09:22:56 am »
I made a an offer for a Advantest R6142 over a year ago on eBay, however the seller declined my offer. The unit has been on sale on eBay for over 3 years. It is sold 3 hours after Dave did a video on it. Haha, gotta love his influence.

 :-+
 

Offline gemby

  • Contributor
  • Posts: 24
  • Country: hr
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #31 on: February 05, 2014, 09:26:30 am »
 

Offline EEVblogTopic starter

  • Administrator
  • *****
  • Posts: 38249
  • Country: au
    • EEVblog
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #32 on: February 05, 2014, 10:01:48 am »
I've got an R6871E on the way...
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13270
  • Country: gb
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #33 on: February 05, 2014, 10:05:11 am »
@Dave,

The R6871E is a sizeable beast but has excellent performance. I own one and IIRC that was the unit for which I imported the user & service manual from the USA. I believe that these manuals are now available for free download.

Update: User Manual is here:

http://www.ko4bb.com/manuals/index.php?dir=Advantest

Service Manual is here:

http://www.eserviceinfo.com/downloadsm/62929/Advantest_R6871E.html
« Last Edit: February 05, 2014, 10:10:30 am by Aurora »
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Slothie

  • Regular Contributor
  • *
  • Posts: 66
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #34 on: February 05, 2014, 05:30:07 pm »
Just looked on eBay UK and there are a few Advantest spectum analysers for sale at 2-3K and voltage/current sources at 1.5-2K. No 300 dollar bargains! But perhaps they are rarer in this neck of the woods.
 

Offline hikariuk

  • Supporter
  • ****
  • Posts: 206
  • Country: gb
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #35 on: February 05, 2014, 07:11:03 pm »
Just looked on eBay UK and there are a few Advantest spectum analysers for sale at 2-3K and voltage/current sources at 1.5-2K. No  bargains! But perhaps they are rarer in this neck of the woods.

Yeah, I noticed that too.  Set up a search anyway, Just In Case :)
I write software.  I'd far rather be doing something else.
 

Offline drtaylor

  • Regular Contributor
  • *
  • Posts: 120
  • Country: us
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #36 on: February 11, 2014, 12:02:13 am »
Dave,. I just wanted to comment on the 2-wire vs 4-wire sensing on the R6142. It is common practice to hook the sense leads to the source leads internally with resistors. These resistors are sized so that the sensing error they add is insignificant which is why you didn't see a change in output value. But if you use 4 terminal sensing at the end of your cable, and there is a load there, you will definitely improve the delivery point accuracy by using kelvin connections. For example, let's say the sense and source leads are connected internally with 200 ohm resistors. Then try to source 100mA at 10 V through 22AWG wire. Without the sense leads, the wire resistance alone will cause a significant error. At Fluke we'd debate leaving out the 2T/4T switch as the resistors add so little error in low current measurements.

Obviously Advantest Engineers went with the resistor connected sense leads, so, for low current outputs, you would not see a difference between the 2T and 4T positions of the switch.
 
The following users thanked this post: Kean

Offline megajocke

  • Regular Contributor
  • *
  • Posts: 191
  • Country: 00
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #37 on: March 01, 2014, 08:58:08 pm »
I'm not sure I understood the video correctly, but to me it sounded like Dave tested flipping the switch in the constant current mode -- which wouldn't even use the 4-wire connection. So nothing changing is the expected behavior in that case.

There are probably resistors as you say, so even in constant voltage mode it would probably still work, albeit possibly with a little bit of voltage error, if switched to 4-wire sensing while leaving the sense terminals unconnected.

To test the 4-wire mode you'd need a load causing significant voltage drop in the cables and also to operate it in CV mode.
« Last Edit: March 01, 2014, 09:00:19 pm by megajocke »
 

Offline Jebnor

  • Regular Contributor
  • *
  • Posts: 170
  • Country: ca
  • Absolutely! Yes, kind of, sort of, not really, no.
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #38 on: August 08, 2014, 02:30:30 am »
Have we had a teardown of this lovely piece of test equipment yet?  Have I missed it?
Before this, there was a typo.
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 17006
  • Country: us
  • DavidH
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #39 on: August 09, 2014, 11:20:26 am »
What's the practical use for this generator other than calibrating other test equipment?

Precision power supplies like this can be used for device characterization and design and development work.  Keithley has a whole business selling precision sources like this.
 

Offline David Hess

  • Super Contributor
  • ***
  • Posts: 17006
  • Country: us
  • DavidH
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #40 on: August 09, 2014, 11:26:58 am »
Dave,. I just wanted to comment on the 2-wire vs 4-wire sensing on the R6142. It is common practice to hook the sense leads to the source leads internally with resistors. These resistors are sized so that the sensing error they add is insignificant which is why you didn't see a change in output value. But if you use 4 terminal sensing at the end of your cable, and there is a load there, you will definitely improve the delivery point accuracy by using kelvin connections. For example, let's say the sense and source leads are connected internally with 200 ohm resistors. Then try to source 100mA at 10 V through 22AWG wire. Without the sense leads, the wire resistance alone will cause a significant error. At Fluke we'd debate leaving out the 2T/4T switch as the resistors add so little error in low current measurements.

Obviously Advantest Engineers went with the resistor connected sense leads, so, for low current outputs, you would not see a difference between the 2T and 4T positions of the switch.

Adding resistors or some type of low voltage shunt is also a safety feature.  If a sense lead accidentally gets disconnected, the force voltage will not rise out of bounds possibly damaging the load.  Power supplies with remote sense often include this to protect their load.
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5517
  • Country: de
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #41 on: June 14, 2015, 09:33:54 pm »
I got a bargain price on a Advantest R6144 Current Voltage Generator in perfect condition
This R6144 is very similar to the one Dave showed in this video.
But it took me some time to find the operating guide / user manual but I finally got lucky.

Advantest R6144 User Manual:
ds.theatr.us/test-equipment/Advantest/6144_ope_e.pdf

If anyone wants to see some pictures from the inside of this R6144, let me know, and I will open it.
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline baoshi

  • Regular Contributor
  • *
  • Posts: 167
  • Country: sg
    • Digital Me
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #42 on: August 06, 2015, 11:53:27 pm »
Thanks for the manual. I heard it is possible to hack R6142 into a R6144 by shorting JP1, JP2 and JP11.  Maybe Dave or you can tear down your units to compare?
 

Offline dadler

  • Supporter
  • ****
  • Posts: 851
  • Country: us
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #43 on: August 07, 2015, 02:16:34 am »
I also have an Advantest R6144, excellent condition, still in perfect calibration. Purchased for peanuts on eBay from a seller that had one left and was liquidating them. Excellent unit.
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5517
  • Country: de
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #44 on: August 07, 2015, 07:43:24 am »
Thanks for the manual. I heard it is possible to hack R6142 into a R6144 by shorting JP1, JP2 and JP11.  Maybe Dave or you can tear down your units to compare?
I am traveling right now, but let me take a look when I am back in the lab.
Also, I got one more manual for this R6144, may be I find the 'hack' possibility in there.
Inreresting really is how stable this unit behaves
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline Len

  • Frequent Contributor
  • **
  • Posts: 551
  • Country: ca
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #45 on: August 07, 2015, 03:21:04 pm »
Dave uses his R6142 to troubleshoot a Keithley DMM in EEVblog #777
https://youtu.be/kdCfAR06GeI?t=8m3s
DIY Eurorack Synth: https://lenp.net/synth/
 

Offline kubatyszko

  • Supporter
  • ****
  • Posts: 8
  • Country: jp
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #46 on: August 16, 2015, 11:45:04 am »
Oh and btw, the TR6142 is the older model. Apart from the datasheet in the video saying so, at some point Advantest changed revisions in their test kit by removing the T at the start - usually the models internals were updated with this change too. I have no idea why the T was dropped off rather than change the model number!

I have found that the schematics of one revision often very similar to the newer model, an element of design reuse. So the schematics in the older TR6142 may line up in some respects.

Si.

For the general public, the "TR" models seem to have been made under brand "Takeda Riken" - and many people in Japan still refer to Advantest as Takeda.

The reason for replacing the TR with R was just the brand change.

Also, there's one more brand called "ADCMT" which seems to hold the "old" equipment, at least Adantest seems to direct queries regarding older equipment there...
 

Offline deadlylover

  • Frequent Contributor
  • **
  • Posts: 325
  • Country: au
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #47 on: August 21, 2015, 03:21:41 pm »
Yep, it seems like any new electronic test gear is now under the ADCMT brand. There are still a few spectrum analysers on the Advantest site, but everything else is ADCMT.

I have both the TR6142 and R6142, they are certainly a little different on the inside and outside. You have to twiddle pots to adjust the TR6142, whereas the R6142 is done err... closed case style in software? (I forget the proper term =P)
Otherwise, most things from "TR" to "R" are the same.

Just a friendly warning, the Maxell backup batteries used inside most old-ish Advantest gear will need to be replaced soon, I had one explode in a nice R6145 Voltage/Current Source/Monitor, taking out quite a bit of the digital circuitry.  :-[
So don't be lazy and replace them ASAP!  :P
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5517
  • Country: de
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #48 on: August 21, 2015, 03:43:18 pm »
Just a friendly warning, the Maxell backup batteries used inside most old-ish Advantest gear will need to be replaced soon, I had one explode in a nice R6145 Voltage/Current Source/Monitor, taking out quite a bit of the digital circuitry.  :-[
So don't be lazy and replace them ASAP!  :P
Thanks for that info
I will take my unit apart and exchange it and at the same time take some pictures.
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline deadlylover

  • Frequent Contributor
  • **
  • Posts: 325
  • Country: au
Re: EEVblog #576 - Advantest R6142 Current Voltage Generator
« Reply #49 on: August 21, 2015, 03:54:33 pm »
Thanks for that info
I will take my unit apart and exchange it and at the same time take some pictures.

I was a bit lazy because I can recalibrate the unit myself any time I wanted, so I thought it was okay if the battery dies a silent death...  :-\

Murphy got me.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf