Hi,
If you look at input sections of the classic Tektronix scopes 465 etc and the 7A18 and 7A26 plugins for the 7000 series. The signal path is very similar.
The input attenuator is arranged in 1, 2, 5 sequence. 10x attenuators are added to the signal path for 10, 20, 50x attenuation.
The basic sensitivity, with the input attenuator set to zero attenuation, in most of the Tektronix scopes, is 10mV per division. For 5mV/div 2x amplification is added to the signal chain.
To obtain 500uV /div a 20x amplifier would be needed.
To maintain a 100 MHz bandwidth, the amplifier needs a GBW product of 2GHz.
Regards,
Jay_Diddy_B