Hi group,
I didn't buy it today, but I got it last week. I obtained a Rode & Schwartz FSH3 Hand Held Spectrum Analyzer from eBay.
The datasheet can be found here:
https://cdn.rohde-schwarz.com/pws/dl_downloads/dl_common_library/dl_brochures_and_datasheets/pdf_1/FSH_dat_en.pdfThere are three models of the FSH3.
The .03 has the preamplifier
The .13 has the tracking generator
The .23 has both the generator and tracking generator.
There is a teardown and repair video (The Signal Path) of another unit here:
I had bought another FSH3.23 and attempted repair. The repair is documented here:
https://www.eevblog.com/forum/testgear/repair-rohde-shwarz-fsh3-portable-spectrum-analyzer/I had limited success with the repair.
1) fixed the 30V power supply, as documented in the thread.
2) I fixed the RF module, this is not documented. But it was a 3.3 Ohm resistor that was open circuit which prevent -5V from getting to the ADC driver.
3) The main board had an issue that the contrast adjustment was broken
4) The LCD module was damaged.
5) The LCD module is khs057qv1cr-g01. The one in the unit has 22 pin connector. I did find some displays with the same part number but they had a 16 pin connector.
I was trolling eBay for a couple of years before a suitable unit showed up.
Here is a picture from the eBay listing:
And some detailed pictures:
One of the previous owners must have damaged the screen. They cut the center out of the screen and replaced just the center.
I asked R&S for the prices on some parts. I got $360.00 USD for the plastic window and $1150.00 for the LCD module
These are approximate conversions from Canadian dollars.
I took the RF section and the case from the from the first unit that I bought and used the main board and display from the eBay unit.
I now have a very nice unit.
I am using the tracking generator to measure the frequency response of a mini-circuits 10.7MHz low pass filter.
Regards,
Jay_Diddy_B