Great, that's what that cal board is for, and luckily having only one pot to adjust will make it easy for you to maintain its accuracy. Old style pot DMM from even Fluke required only a few pot adjustments to do the cal; today you have to set a constant for every range, which is more work. Luckily they don't go out of cal often, if at all.
The Uni-T should hold its cal if the parts used are good quality. There are many reasons it could still drift, an unaged internal reference is one, and temperature stresses is another common one. DMM shouldn't drift much with humidity and barometric pressure as the variation is often in the uV, but with lesser DMMs you never know. The biggest drift from internal reference aging occurs in the 1st 200 hours, and its prudent to preage internal references for use if you want to minimize this phenomenon. For temperature stress, put the DMM in a refrigerator for ~ 40F exposure and summer sunlight to your regular temperature, typically 100-120F. Just cover the unit so UV light won't accelerate wear on the the plastic or the paint labels. If the unit is rated to those temps it should still read within spec and when it cools down to room temp it should not have drifted from your initial adjustment. There are many reasons thereafter for what parts could be suboptimal to cause the drift, a common one being the adjustable pot, the wiper may nudge just a little, contracting with cold, and expanding with heat, but when back at room temp it should be at the baseline setting.
The voltagestandard.com board seems to hold its factory cal very well. There are no formal studies just anecodotes from time-nuts, volt-nuts and the hp_agilent_equipment yahoogroup; the calibrated output has remained stable >= 3 years, IIRC. You can find comments about them in the archives.
I have a Geller SVR board that is now just 1 year old, its holding its original cal of 10.000 00 VDC very well.
I ordered and received a DMM Check from Doug at http://www.voltagestandard.com. I was finally able to check and adjust my UT71E. It was reading low on the 40 scale so I adjusted it to get a 5.000 reading with the DMM Check. I then checked the meter against my reference meters and it seems that the base adjust VR2 affects all ranges as well. So I appear to have an accurate meter. I am awaiting for some more reference equipment to confirm the the results.
The meter seems to read correctly now, but for how long?