alm, or PedroDaGr8,
So, something still isn't quite right. I did replace the inverter, and that seems to be working now. But...
On the 30V and 300V range, with no voltage applied to the inputs, or the inputs not shorted, the voltage is pretty stable at 00.0000 or whatever. But on the 3V or 300mV range, if there's no voltage applied, or the inputs aren't shorted, the reading creeps up (down) toward -1.5V or more. On the 300mV range, this relatively quickly leads to an overflow, but on the 3V range, it just gradually creeps. If I hook anything up to the inputs, like my FG, even with the output off, the impedance is low enough for the creep to not happen or be much slower.
Sounds like some type of capacitance on the inputs?
Do either of your working meters do anything like this? I know some drift is common from it picking up transient voltages... but to pick up over a volt...
I will say that it reads accurately when I give it a voltage, and it zeros properly when I short the leads together...
Thoughts? And thanks!