Interesting picture.
Cold Cathode X-Ray tube with absolutely no shielding (not a great idea), Low resolution scintillator plate and a digital SLR with no X-Ray screening glass to protect the CCD chip.
Not a setup I would recommend to anyone..... why ?
1. That cold cathode tube was designed to sit inside the EHT transformer that was cooled with oil. The Oil acts as a coolant and X-Ray reduction device.
2. The scintillator plate is an image enhancer (intensifier screen) from a film type X-Ray plate. Quality and sensitivity can vary greatly. Resolution is usually not that great when used with a digital camera in the shown format..
3. The CCD imaging chip and associated electronics may be adversely effected by the X-Ray energy hitting them. This causes spotting on the image. If a CCD chip is in use, the exposure to X-Ray can permanently increase its dark current, causing the camera harm. It is better to use a 45 Degree mirror and image the intensifier screen with the camera out of the main X-Ray energy beam.
That is very much a Heath Robinson X-Ray imaging setup that should only be used with great experience of X-Ray energy and appropriate precautions. It would seriously fail the HSE regulations in the UK on many counts
It stands as an example of why real X-Ray equipment and tubes are licenced in some countries. In the UK we are a little more open minded though.
I became a licenced radiographer through my work and was suitably Harwell 'badged' for many years. I never had an abnormal dose reading in my career. Correctly configured equipment and use is the key to safe X-Ray imaging. Soft X-Ray is easily attenuated, but still harmful at high accumulated doses.
Like many things in life, X-Ray may be used safely, but abusers of such need to realise that damage to the skin, or worse can result.
Fraser