To bad I missed this topic until now. Good to know I'm not alone in the quest for the holy grale of precision ( but sometimes it feels like chasing windmills)
I had the same troubles about what cables. I tried massive copper ( VD wire, you use for mains in the walls), coax, triax, shieded telephone wire. With normal bannas, goldplated ones, bnc-banana adapters, home made copper forks to screw under terminals ect,
Problem is you get different results but what is the right one ( the man with to many clocks story...)
I am very interested do some tests too when my lab is finnished. I sold a bunch of Racals that were in a 19 " rack. Now all my calibrators and KV deviders are in there and the Guildline standard cell cabinet is on top.
I have also the 1010 resistors and a GR decade. I can meaure G for high Ohms
I can measure using a GR1608, Boonton 63H, Prema and solartron 7,5 digit, Keithley 2000 and 196.
Also a HP sourcemeter that goes into the PetaOhm region but it is analog.
And the two KV deviders ( Fluke 720 and ESI decavider) but I have to figure out how to do this. There is a brief description in the Fluke manual as I recall.
Maybe there is a way to measure the one KV with the other using a calibration source and a null detector and so calculating some parmeters ?
Problems with seebeck I only experienced while calibrating, but if you wait long enough the contacts have the same temperature so no seebeck ( I assume, Roberts seems to tell the same if I understand it correct)
I had problems with dirt ( for very high Ohm) , pressure ( the grip of the kelvin clips, but even tiurning a banana a bit can give different reading in the mOhm or lower end) and current that heats up the resistor.
I found out that cleaning a 10 M resistor with ipa and steelwool for the legs made the difference between unstable and stable readings, but the writer of my Bible ( analog SEEKrets) seems to have other experiences. I measured leakage of my Fluke 720. I am very bad in remembering numbers but it was a factor 100 to much leakage. Putting the KV on 2 cleaned teflon blocks made leakage less, so the table conducted to much. i then cleaned the whole inside and outside with TRI ( at that time it looked like a good idea, and IPA after that) that improved leakage about 150X. Still not what the factory described but the best I could do. ( my 720 was total loss when I got it, had to make all swtich isolation myself, because all the studs and wiper adapters were broken in many peaces. Photos are on my site)
Now I stop, I can talk about this stuff for hours ;-)
Should i use some format for testing. You talk about a spreadsheet but I have to rest now so I'm typing this on my ipad and I can not open it right now, but should i use that ?
The Keithley book is for me too the source of this madness :-)