Hi! I also plan to replace this FAN. There are several models from 170CFM to 250CFM. Did you figure out flow of the original FAN?
Yes it's 180 CFM
I would just get something of similar wattage, 22W was mentioned above. I doubt its that critical.
It can be on these older HP instruments, 3335A, 8340B, 8663A etc. They rely on positive case pressure to force air around inside the unit and through small holes in the RF blocks. Static pressure of the fan is also important, the fan has got to be able to generate some pressure to force the air through.
That said, the fans were probably specced with worst case scenario in mind, running 24/7 in a tightly packed rig at 55deg (C) ambient. There probably is some leeway when fitting a replacement for light use at home.
In this case I fitted a 130 CFM fan (never did find out the static pressure spec of the OEM one) and everything seems ok so far.