OK folks bad news.
I've run through all the codes on the two links Dave played with.
12-18,28-88 and 04 give clock errors.
Some codes show offsets on the Y traces, and also time offsets between trigger event and trigger point on screen.
03 is the only code that gives 200MHz and apparently plausible operation, but it does have problems.
DC offsets on some Y ranges - difefrent between channels - possibly calibration related?
Enabling serial bus decode prevents Ch2 trigger working ( possibly offset related) .
I tried a user cal and it thought there were still signals connected.
Could be there is some scope by tweaking other links but I'm done with this - I'll leave it to others to play some more.
I returned to factory 24 config and did a user cal. I Just happened to still have terminal open when doing user cal and saw this :
Performing USER calibraion
######################## Channel 1 ########################
**** Baldwin Interpolator ****
fiPreDelay = 0x10
fiCalResult = 180
fiWide = 316
Final Cal Factor = 0x10
**** CAL PASSED **** Time: 1 seconds
**** Talon Calibrator ****
Calibrate AC Signal
Channel 1: Offset 0x20400 GainIndex 7 GainVern 0x1db0
Channel 2: Offset 0x20400 GainIndex 7 GainVern 0x1db0
Calibrate DC Signal
# Pts (out of 20): 20 B1 = 173657.888115 B0 = 126614.320739
# Pts (out of 20): 20 B1 = 170768.501484 B0 = 124845.623921
Testing...
Vin = -0.200000 AvgQLvl = 29.000000 VinActual = -0.193359
Vin = 0.000000 AvgQLvl = 131.500000 VinActual = 0.006836
Vin = 0.200000 AvgQLvl = 234.500000 VinActual = 0.208008
**** CAL PASSED **** Time: 2 seconds
**** Talon ****
**Talon Init Cal**
Slice Offset = 20
Slice offset correction successful
Using all slices for calibration
**Talon Dac Cal**
Gain Dac Iteration 1: Deltas 10 | Max 1 | Min -1
Gain Dac Iteration 2: Deltas 9 | Max 1 | Min -1
Gain Dac Iteration 3: Deltas 8 | Max 1 | Min -1
Offset Dac Iteration 1: Deltas 37 | Max 3 | Min -3
Offset Dac Iteration 2: Deltas 3 | Max 3 | Min -3
**Talon Pipeline Cal**
*** Talon pipeline cal ***
# pts not used = 0
Offset Adjustment = -25.965942
Gain Adjustment = 1.009946
**Talon4 Timing Cal**
Freq: 39.999963MHz
QMax 0xbd QMin 0x3c
Min Per: 2.490246e-008 Max Per: 2.509196e-008
Harmonic Used: 1
MajorPass 1 | Deltas 11 | Max Dac 4 | Min Dac -4
MajorPass 2 | Deltas 5 | Max Dac 5 | Min Dac -5
MajorPass 3 | Deltas 4 | Max Dac 4 | Min Dac -5
MinorPass 1 | Deltas 177 | Max Dac 25 | Min Dac -26
Minor Timing Vernier is clipped
MinorPass 2 | Deltas 80 | Max Dac 34 | Min Dac -37
# pts not used = 0
Offset Adjustment = -27.887817
Gain Adjustment = 1.010676
Using partial slices for calibration
**Talon Dac Cal**
Gain Dac Iteration 1: Deltas 0 | Max 1 | Min 0
Gain Dac Iteration 2: Deltas 0 | Max 1 | Min 0
Gain Dac Iteration 3: Deltas 0 | Max 1 | Min 0
Offset Dac Iteration 1: Deltas 0 | Max 3 | Min -2
Offset Dac Iteration 2: Deltas 0 | Max 3 | Min -2
# pts not used = 0
**** CAL PASSED **** Time: 112 seconds
**** DC Gain ****
28000, 0.812299 : -4418816.035018, 3617399.771942
28400, 0.812208 : -25035.786492 , 48734.277620
28800, 0.796231 : -4304.740688 , 32227.569411
29200, 0.703311 : -4761.655513 , 32548.922414
29600, 0.619306 : -4951.054696 , 32666.218487
30000, 0.538515 : -6040.760514 , 33253.041695
30400, 0.472298 : -6659.858803 , 33545.440844
30800, 0.412237 : -7267.789993 , 33796.052632
31200, 0.357200 : -9472.275061 , 34583.494105
31600, 0.314971 : -9543.879732 , 34606.047516
32000, 0.273060 : -11894.525749 , 35247.913863
32400, 0.239431 : -13151.239035 , 35548.809524
32800, 0.209015 : -14370.134360 , 35803.577236
33200, 0.181180 : -18566.455108 , 36563.865546
33600, 0.159636 : -18803.473684 , 36601.702128
34000, 0.138363 : -22895.421325 , 37167.875648
34400, 0.120892 : -25916.811236 , 37533.137830
34800, 0.105458 : -27704.177528 , 37721.630094
35200, 0.091020 : -36823.469298 , 38551.666667
35600, 0.080157 : -37929.753784 , 38640.343348
36000, 0.069611 : -45089.962406 , 39138.775510
36400, 0.060740 : -59312.970682 , 40002.684564
36800, 0.053996 : -285084.923599, 52193.548387
37200, 0.052593 : 0.000000 , 0.000000
Reference range = 0.472V
Coarse gain ratio
0.1x = 0.357 (1.324V)
0.4x = 1.000 (0.472V)
0.8x = 2.828 (0.167V)
2.0x = 7.846 (0.060V)
Gain# MIN MAX %OL
0 0.006703 0.103532
1 0.018598 0.287243
2 0.052593 0.812299
3 0.147422 2.276928
4 0.302873 4.677844
5 0.840301 12.978381
6 2.376298 36.701735
7 6.660921 102.877407
ADC half speed gain ratio = 0.991 (0.477V)
**** CAL PASSED **** Time: 15 seconds
**** Offset Error ****
AC Coupling
Gain# 0 [29487 (0.006703V) - 37407 (0.080000V), 45pts]
Gain# 1 [29614 (0.080000V) - 32629 (0.216000V), 45pts]
Gain# 2 [29559 (0.216000V) - 32754 (0.624000V), 45pts]
Gain# 3 [29826 (0.624000V) - 32661 (1.600000V), 45pts]
Gain# 4 [29617 (1.600000V) - 32002 (3.520000V), 45pts]
Gain# 5 [29677 (3.520000V) - 32692 (9.600000V), 45pts]
Gain# 6 [29569 (9.600000V) - 32809 (28.000000V), 45pts]
Gain# 7 [29497 (28.000000V) - 32692 (80.500000V), 45pts]
DC Coupling
AC Coupling
DC Coupling
0 retry out of 920 offset searched
**** CAL PASSED **** Time: 21 seconds
**** Trigger Level ****
div -3, pos slope dac = 0x95C3, neg slope dac = 0x94FD, avg = 38240.000
div -2, pos slope dac = 0x9341, neg slope dac = 0x9275, avg = 37595.000
div -1, pos slope dac = 0x90BB, neg slope dac = 0x8FF3, avg = 36951.000
div 0, pos slope dac = 0x8E37, neg slope dac = 0x8D6F, avg = 36307.000
div 1, pos slope dac = 0x8BB5, neg slope dac = 0x8AE7, avg = 35662.000
div 2, pos slope dac = 0x8931, neg slope dac = 0x8869, avg = 35021.000
div 3, pos slope dac = 0x86AB, neg slope dac = 0x85E3, avg = 34375.000
Trig 1 B1 = -644.000, B0 = 36307.286
div -3, pos slope dac = 0x959D, neg slope dac = 0x94DF, avg = 38206.000
div -2, pos slope dac = 0x9317, neg slope dac = 0x925B, avg = 37561.000
div -1, pos slope dac = 0x9095, neg slope dac = 0x8FD7, avg = 36918.000
div 0, pos slope dac = 0x8E0D, neg slope dac = 0x8D4F, avg = 36270.000
div 1, pos slope dac = 0x8B89, neg slope dac = 0x8ACB, avg = 35626.000
div 2, pos slope dac = 0x8905, neg slope dac = 0x8843, avg = 34980.000
div 3, pos slope dac = 0x8681, neg slope dac = 0x85C3, avg = 34338.000
Trig 2 B1 = -644.929, B0 = 36271.286
div -3, pos slope dac = 0x959B, neg slope dac = 0x94E1, avg = 38206.000
div -2, pos slope dac = 0x9315, neg slope dac = 0x925D, avg = 37561.000
div -1, pos slope dac = 0x9091, neg slope dac = 0x8FD9, avg = 36917.000
div 0, pos slope dac = 0x8E0B, neg slope dac = 0x8D4F, avg = 36269.000
div 2, pos slope dac = 0x8903, neg slope dac = 0x8847, avg = 34981.000
div 3, pos slope dac = 0x867F, neg slope dac = 0x85BF, avg = 34335.000
Trig 3 B1 = -645.143, B0 = 36270.714
Trig 1 - AC offset -47.285714
Trig 2 - AC offset -46.285714
Trig 1 - LF Rej offset -46.285714
Trig 2 - LF Rej offset -46.285714
Trig 1 - HF_Rej offset -1.285714
Trig 2 - HF_Rej offset -1.285714
Trig 1 - AC_HF_REJ offset -47.285714
Trig 2 - AC_HF_REJ offset -46.285714
**** CAL PASSED **** Time: 29 seconds
**** Trigger Hysteresis ****
Hyst dac = 12288, pos slope dac = 0x8E27, neg slope dac = 0x8D79, range = 174, div = 0.270 +
Hyst dac = 14336, pos slope dac = 0x8E49, neg slope dac = 0x8D5B, range = 238, div = 0.370 +
Hyst dac = 16384, pos slope dac = 0x8E67, neg slope dac = 0x8D39, range = 302, div = 0.469 +
Hyst dac = 18432, pos slope dac = 0x8E87, neg slope dac = 0x8D17, range = 368, div = 0.571 +
Hyst dac = 20480, pos slope dac = 0x8EA9, neg slope dac = 0x8CF9, range = 432, div = 0.671 +
Hyst dac = 22528, pos slope dac = 0x8ECB, neg slope dac = 0x8CD5, range = 502, div = 0.780 +
Hyst dac = 24576, pos slope dac = 0x8EED, neg slope dac = 0x8CB9, range = 564, div = 0.876 +
Hyst dac = 26624, pos slope dac = 0x8F0D, neg slope dac = 0x8C8F, range = 638, div = 0.991 +
Hyst dac = 28672, pos slope dac = 0x8F33, neg slope dac = 0x8C71, range = 706, div = 1.096 +
Hyst dac = 30720, pos slope dac = 0x8F57, neg slope dac = 0x8C4D, range = 778, div = 1.208 +
Hyst dac = 32768, pos slope dac = 0x8F7D, neg slope dac = 0x8C25, range = 856, div = 1.329 +
Hyst dac = 34816, pos slope dac = 0x8F9F, neg slope dac = 0x8BFF, range = 928, div = 1.441 +
Hyst dac = 36864, pos slope dac = 0x8FC7, neg slope dac = 0x8BD9, range = 1006, div = 1.562 +
Trig Hyst B1 = 19074.498, B0 = 7506.466
**** CAL PASSED **** Time: 13 seconds
**** External Trigger Level ****
/1 Trig B1 = 5303.000, B0 = 35846.000
/5 Trig B1 = 1071.000, B0 = 35846.000
/1 Trig 2 B1 = 17283.000, B0 = 28351.000
/5 Trig 2 B1 = 3492.000, B0 = 28351.000
**** Wave Gen ****
[calibrator] = 0x3fb22 Q | -52958.38 Q/V
[0: DIV1 + DIV1 ] = 0x1e4f0 Q | 10141.20 Q/V | (calV=2.490125V) 33.965306 V
[1: DIV4 + DIV1 ] = 0x1f948 Q | 10141.60 Q/V | (calV=2.499812V) 8.412196 V
[2: DIV16 + DIV1 ] = 0x1fde1 Q | 10140.40 Q/V | (calV=0.640975V) 2.122063 V
[3: DIV64 + DIV1 ] = 0x1fec1 Q | 10141.20 Q/V | (calV=0.159956V) 0.547273 V
[4: DIV1 + DIV23] = 0x1e4f4 Q | 236185.00 Q/V | (calV=0.451354V) 1.447447 V
[5: DIV4 + DIV23] = 0x1f941 Q | 236197.50 Q/V | (calV=0.109916V) 0.361558 V
[6: DIV16 + DIV23] = 0x1fdd5 Q | 236200.00 Q/V | (calV=0.024963V) 0.090880 V
[7: DIV64 + DIV23] = 0x1feb7 Q | 236202.50 Q/V | (calV=0.006987V) 0.023434 V
[0: DIV1 + DIV1 ] = 33.677039 V (overwrite)
**** CAL PASSED **** Time: 62 seconds
**** Wavegen Trigger Level Offset ****
Wavegen Trig Offset = 0x8C11
**** Baldwin Trig Time Qual ****
Set CalConfigScope range 1.000000E-005, delay -1.000000E-006
Top : 1.859297
Base : -29.497488
Vert Scale : 80.000000
Vert Offset: -14.748744
Trig Level : -14.748744
Horz Range : 1.000000e-007
Horz Delay : 1.500000e-008
CalConfigScope range 1.000000E-007, delay 1.500000E-008
--- Internal Timer ---
tVolt for Edge Trigger = 1.722706 ns
Zero Time = 1.722706 ns
*Int Osc Freq*
Oscillator Freq: 368.472722 MHz
Set CalConfigScope range 5.000000E-008, delay -5.000000E-009
CalConfigScope range 5.000000E-008, delay 1.000000E-008
*Int Find Min Time*
Zero Time Timer 0 : 1348.106337 ps
Zero Time Timer 1 : 1308.042235 ps
*Int Find Min Time*
Timer 0:
Dac 0 Time 1366.135183 ps
Dac 31 Time 2501.899775 ps
Dac 62 Time 3637.720940 ps
Dac 93 Time 4759.564671 ps
Dac 124 Time 5929.436465 ps
Fine Dac Gain 36.723443 ps/DacCode
Timer 1:
Dac 0 Time 1338.248459 ps
Dac 31 Time 2500.144721 ps
Dac 62 Time 3620.320828 ps
Dac 93 Time 4759.815071 ps
Dac 124 Time 5904.346321 ps
Fine Dac Gain 36.747955 ps/DacCode
*Int Fast Path*
Fast Dac Zero Time 951.735301 ps
Dac 0 Time 978.348057 ps
Dac 31 Time 1767.935959 ps
Dac 62 Time 2945.978723 ps
Dac 93 Time 4102.513389 ps
Dac 124 Time 5264.372363 ps
Fine Dac Gain 35.182665 ps/DacCode
--- Sync Timer ---
Set CalConfigScope range 5.000000E-007, delay -5.000000E-008
MClk Frequency = 200006501.105188
*Calibrating PdSel*
PdSel 0 : Min 62 Max 117
PdSel = 0
Min Fine Dac = 133
*Fine Dac*
Sync Timer minimum Time : 2.183431e-008
Dac 133 Val: -2.177687e-008
Dac 227 Val: -2.624178e-008
Dac 321 Val: -3.057086e-008
Dac 415 Val: -3.495388e-008
Dac 509 Val: -3.939129e-008
Fine Dac Gain: 2.139190e+010 DacCode/second
Fine Dac Gain: 46.746673 ps/DacCode
**** Baldwin Setup and Hold ****
--- Vertical Settings ---
Top : 1.934667
Base : -29.422117
Vert Offset: -14.711058
Trig Level : -14.711058
*Clk Path*
Long Delay Dac 0 Time 1116.976351 ps
Long Delay Dac 31 Time -8571.428571 ps
Long Delay Dac 62 Time -18325.591216 ps
Long Delay Dac 93 Time -28122.970779 ps
Long Delay Dac 124 Time -37865.953947 ps
Zero Pt with long delay : 1149.686928 ps
Long Delay Gain : 314.572267 ps / daccode
Trombone Delay Gain : 73.889024 ps / daccode
*Data Path*
Min TVolt = -8437.500000ps
Max TVolt = -2187.500000ps
DataPath Offset = 1250.000000ps
ClockPath Offset = 3337.186928ps
**** CAL PASSED **** Time: 3 seconds
**** Baldwin Pattern Trig ****
**Pattern Delay Cal**
Top : 1.972417
Base : -29.384367
Vert Offset: -14.692184
Trig Level : -14.692184
Zero Time = 1.808594ns
DeltaT = 1 : Time Delta = 3.028106ns
DeltaT = 2 : Time Delta = 6.002006ns
DeltaT = 3 : Time Delta = 9.006373ns
DeltaT Gain : 2.999302ns/code
DelatT Offset: 0.010168ns
TBone = 1 : Time Delta = 0.100123ns
TBone = 9 : Time Delta = 0.716955ns
TBone = 17 : Time Delta = 1.277744ns
TBone = 25 : Time Delta = 1.859992ns
TBone = 33 : Time Delta = 2.466146ns
TBone = 41 : Time Delta = 3.054371ns
TBone = 49 : Time Delta = 3.621916ns
TBone = 57 : Time Delta = 4.172903ns
TBone Gain : 2.999302ns/code
TBone Offset: 0.010168ns
Compensation of 1.750000ns:
DeltaT: 0
Tbone : 23
**Xtrig Delay Cal**
Final Qual Delay Setting: 5
######################## Channel 2 ########################
**** DC Gain ****
28400, 0.797058 : -25330.306767 , 48589.714286
28800, 0.781266 : -4127.377732 , 32024.581006
29200, 0.684352 : -4787.045015 , 32476.025918
29600, 0.600794 : -5321.493018 , 32797.118848
30000, 0.525627 : -6043.358806 , 33176.550784
30400, 0.459438 : -7115.254710 , 33669.020867
30800, 0.403221 : -7662.581995 , 33889.714780
31200, 0.351019 : -9293.089485 , 34462.054507
31600, 0.307977 : -10178.653695 , 34734.787600
32000, 0.268679 : -11665.272853 , 35134.210526
32400, 0.234389 : -13702.638900 , 35611.746522
32800, 0.205197 : -14875.180148 , 35852.348993
33200, 0.178307 : -18279.602821 , 36459.381443
33600, 0.156425 : -20287.431049 , 36773.455378
34000, 0.136708 : -22732.326586 , 37107.692308
34400, 0.119112 : -26947.134858 , 37609.726444
34800, 0.104268 : -28415.408232 , 37762.820513
35200, 0.090191 : -36634.741192 , 38504.132231
35600, 0.079273 : -39935.168326 , 38765.765766
36000, 0.069256 : -43246.865212 , 38995.121951
36400, 0.060007 : -57945.146199 , 39877.124183
36800, 0.053104 : -277050.230264, 51512.500000
37200, 0.051660 : 0.000000 , 0.000000
Reference range = 0.459V
Coarse gain ratio
0.1x = 0.357 (1.287V)
0.4x = 1.000 (0.459V)
0.8x = 2.824 (0.163V)
2.0x = 7.791 (0.059V)
Gain# MIN MAX %OL
0 0.006631 0.102303
1 0.018294 0.282254
2 0.051660 0.797058
3 0.144698 2.232513
4 0.299557 4.621817
5 0.826476 12.751548
6 2.333889 36.009154
7 6.537089 100.859579
ADC half speed gain ratio = 0.990 (0.464V)
AC Coupling
Gain# 0 [29409 (0.006631V) - 37419 (0.080000V), 45pts]
Gain# 1 [29544 (0.080000V) - 32559 (0.216000V), 45pts]
Gain# 2 [29465 (0.216000V) - 32705 (0.624000V), 45pts]
Gain# 3 [29733 (0.624000V) - 32613 (1.600000V), 45pts]
Gain# 4 [29547 (1.600000V) - 31977 (3.520000V), 45pts]
Gain# 5 [29585 (3.520000V) - 32645 (9.600000V), 45pts]
Gain# 6 [29476 (9.600000V) - 32761 (28.000000V), 45pts]
Gain# 7 [29400 (28.000000V) - 32640 (80.500000V), 45pts]
DC Coupling
AC Coupling
DC Coupling
div -3, pos slope dac = 0x959B, neg slope dac = 0x94D1, avg = 38198.000
div -2, pos slope dac = 0x9315, neg slope dac = 0x924B, avg = 37552.000
div -1, pos slope dac = 0x9093, neg slope dac = 0x8FC9, avg = 36910.000
div 0, pos slope dac = 0x8E0F, neg slope dac = 0x8D47, avg = 36267.000
div 1, pos slope dac = 0x8B8D, neg slope dac = 0x8ABF, avg = 35622.000
div 2, pos slope dac = 0x890B, neg slope dac = 0x883F, avg = 34981.000
div 3, pos slope dac = 0x8683, neg slope dac = 0x85BB, avg = 34335.000
Trig 1 B1 = -643.536, B0 = 36266.429
div -3, pos slope dac = 0x9573, neg slope dac = 0x94B3, avg = 38163.000
div -2, pos slope dac = 0x92F1, neg slope dac = 0x922F, avg = 37520.000
div -1, pos slope dac = 0x906D, neg slope dac = 0x8FA9, avg = 36875.000
div 0, pos slope dac = 0x8DE3, neg slope dac = 0x8D25, avg = 36228.000
div 1, pos slope dac = 0x8B63, neg slope dac = 0x8AA1, avg = 35586.000
div 2, pos slope dac = 0x88DD, neg slope dac = 0x881D, avg = 34941.000
div 3, pos slope dac = 0x8659, neg slope dac = 0x8597, avg = 34296.000
Trig 2 B1 = -644.571, B0 = 36229.857
div -3, pos slope dac = 0x95F1, neg slope dac = 0x9539, avg = 38293.000
div -2, pos slope dac = 0x936D, neg slope dac = 0x92B7, avg = 37650.000
div -1, pos slope dac = 0x90E9, neg slope dac = 0x902B, avg = 37002.000
div 0, pos slope dac = 0x8E5F, neg slope dac = 0x8DA3, avg = 36353.000
div 1, pos slope dac = 0x8BD7, neg slope dac = 0x8B1F, avg = 35707.000
div 2, pos slope dac = 0x8951, neg slope dac = 0x8897, avg = 35060.000
div 3, pos slope dac = 0x86CD, neg slope dac = 0x8611, avg = 34415.000
Trig 3 B1 = -646.750, B0 = 36354.286
Trig 1 - AC offset -5.428571
Trig 2 - AC offset -5.857143
Trig 1 - LF Rej offset -5.428571
Trig 2 - LF Rej offset -3.857143
Trig 1 - HF_Rej offset -3.428571
Trig 2 - HF_Rej offset -1.857143
Trig 1 - AC_HF_REJ offset -5.428571
Trig 2 - AC_HF_REJ offset -3.857143
Hyst dac = 12288, pos slope dac = 0x8E01, neg slope dac = 0x8D51, range = 176, div = 0.273 +
Hyst dac = 14336, pos slope dac = 0x8E21, neg slope dac = 0x8D2F, range = 242, div = 0.376 +
Hyst dac = 16384, pos slope dac = 0x8E3F, neg slope dac = 0x8D11, range = 302, div = 0.469 +
Hyst dac = 18432, pos slope dac = 0x8E61, neg slope dac = 0x8CF1, range = 368, div = 0.572 +
Hyst dac = 20480, pos slope dac = 0x8E7F, neg slope dac = 0x8CCD, range = 434, div = 0.674 +
Hyst dac = 22528, pos slope dac = 0x8EA1, neg slope dac = 0x8CAF, range = 498, div = 0.774 +
Hyst dac = 24576, pos slope dac = 0x8EC5, neg slope dac = 0x8C89, range = 572, div = 0.889 +
Hyst dac = 26624, pos slope dac = 0x8EE7, neg slope dac = 0x8C69, range = 638, div = 0.991 +
Hyst dac = 28672, pos slope dac = 0x8F0B, neg slope dac = 0x8C47, range = 708, div = 1.100 +
Hyst dac = 30720, pos slope dac = 0x8F2F, neg slope dac = 0x8C21, range = 782, div = 1.215 +
Hyst dac = 32768, pos slope dac = 0x8F51, neg slope dac = 0x8BFF, range = 850, div = 1.321 +
Hyst dac = 34816, pos slope dac = 0x8F79, neg slope dac = 0x8BD9, range = 928, div = 1.442 +
Hyst dac = 36864, pos slope dac = 0x8FA1, neg slope dac = 0x8BB3, range = 1006, div = 1.563 +
Trig Hyst B1 = 19122.640, B0 = 7423.609
Hyst dac = 12288, pos slope dac = 0x8E51, neg slope dac = 0x8DAD, range = 164, div = 0.254 +
Hyst dac = 14336, pos slope dac = 0x8E71, neg slope dac = 0x8D93, range = 222, div = 0.343 +
Hyst dac = 16384, pos slope dac = 0x8E8D, neg slope dac = 0x8D75, range = 280, div = 0.433 +
Hyst dac = 18432, pos slope dac = 0x8EAD, neg slope dac = 0x8D57, range = 342, div = 0.529 +
Hyst dac = 20480, pos slope dac = 0x8ECD, neg slope dac = 0x8D2F, range = 414, div = 0.640 +
Hyst dac = 22528, pos slope dac = 0x8EF1, neg slope dac = 0x8D13, range = 478, div = 0.739 +
Hyst dac = 24576, pos slope dac = 0x8F15, neg slope dac = 0x8CF5, range = 544, div = 0.841 +
Hyst dac = 26624, pos slope dac = 0x8F33, neg slope dac = 0x8CCB, range = 616, div = 0.952 +
Hyst dac = 28672, pos slope dac = 0x8F57, neg slope dac = 0x8CA7, range = 688, div = 1.064 +
Hyst dac = 30720, pos slope dac = 0x8F79, neg slope dac = 0x8C87, range = 754, div = 1.166 +
Hyst dac = 32768, pos slope dac = 0x8F9F, neg slope dac = 0x8C65, range = 826, div = 1.277 +
Hyst dac = 34816, pos slope dac = 0x8FBF, neg slope dac = 0x8C3F, range = 896, div = 1.385 +
Hyst dac = 36864, pos slope dac = 0x8FE9, neg slope dac = 0x8C19, range = 976, div = 1.509 +
Trig Hyst (2) B1 = 19258.151, B0 = 7693.144
**** CAL PASSED **** Time: 26 seconds
**** Channel Delays ****
Top : 0.251256
Base : -29.497488
Final Scale : 80.000000
Final Offset (POS_SLOPE): -14.748744
Final Offset (NEG_SLOPE): -14.748744
edge trigger delay factors
iteration: 0
analog trigger channel : 1
pos polarity path delay adjustment (Sec) : 1.76809211e-009
pos edge, bi edge delay factor (312.5e-15 Sec counts): -3342
neg edge delay factor (312.5e-15 Sec counts): -3342
analog trigger channel : 2
pos polarity path delay adjustment (Sec) : 1.00390625e-009
pos edge, bi edge delay factor (312.5e-15 Sec counts): -5787
neg edge delay factor (312.5e-15 Sec counts): -5787
pos polarity path delay adjustment (Sec) : 6.2099359e-010
Chan2 TC2 pos edge, bi edge delay factor (312.5e-15 Sec counts): -9513
Chan2 TC2 neg edge delay factor (312.5e-15 Sec counts): -9513
external trigger channel
pos edge, bi edge delay factor (312.5e-15 Sec counts): -595
neg edge delay factor (312.5e-15 Sec counts): -595
iteration: 1
pos polarity path delay adjustment (Sec) : -6.16776316e-012
pos edge, bi edge delay factor (312.5e-15 Sec counts): -3361
neg edge delay factor (312.5e-15 Sec counts): -3361
pos polarity path delay adjustment (Sec) : -1.6025641e-011
pos edge, bi edge delay factor (312.5e-15 Sec counts): -5837
neg edge delay factor (312.5e-15 Sec counts): -5837
pos polarity path delay adjustment (Sec) : 2.00320513e-012
Chan2 TC2 pos edge, bi edge delay factor (312.5e-15 Sec counts): -9507
Chan2 TC2 neg edge delay factor (312.5e-15 Sec counts): -9507
pos edge, bi edge delay factor (312.5e-15 Sec counts): -630
neg edge delay factor (312.5e-15 Sec counts): -630
timeout trigger delay factors
iteration: 0
pos polarity path delay adjustment (Sec) : 3.03285256e-009
pos edge, bi edge delay factor (312.5e-15 Sec counts): -2495
neg edge delay factor (312.5e-15 Sec counts): -2495
pos polarity path delay adjustment (Sec) : 8.05288462e-010
pos edge, bi edge delay factor (312.5e-15 Sec counts): -9623
neg edge delay factor (312.5e-15 Sec counts): -9623
pos edge, bi edge delay factor (312.5e-15 Sec counts): -6058
neg edge delay factor (312.5e-15 Sec counts): -6058
iteration: 1
pos polarity path delay adjustment (Sec) : 0
pos polarity path delay adjustment (Sec) : -2.00320513e-012
pos edge, bi edge delay factor (312.5e-15 Sec counts): -9628
neg edge delay factor (312.5e-15 Sec counts): -9628
pos edge, bi edge delay factor (312.5e-15 Sec counts): -6061
neg edge delay factor (312.5e-15 Sec counts): -6061
setup & hold trigger delay factors
iteration: 0
S&H Delay, pos slope (Sec): 1.15748355e-009
S&H Delay, pos slope factor (312.5e-15 Sec counts): -4796
S&H Delay, neg slope factor (312.5e-15 Sec counts): -4796
iteration: 1
S&H Delay, pos slope (Sec): 2.00320513e-012
S&H Delay, pos slope factor (312.5e-15 Sec counts): -4790
S&H Delay, neg slope factor (312.5e-15 Sec counts): -4790
multi-channel delay factors
iteration: 0
Ch 1 Pattern Delay, pos slope (Sec): 1.07524671e-009
(chan1PattDly[POS_SLOPE]) Ch 1 Pattern Delay, pos slope factor (312.5e-15 Sec counts): -2959
(chan1PattDly[NEG_SLOPE]) Ch 1 Pattern Delay, neg slope factor (312.5e-15 Sec counts): -2959
Ch 1 Pattern Delay, mixed, pos slope delay (Sec): 1.19391026e-009
(chan1PattDlyMixed[POS_SLOPE]) Ch 1 Pattern Delay, mixed, pos slope factor (312.5e-15 Sec counts): -8179
(chan1PattDlyMixed[NEG_SLOPE]) Ch 1 Pattern Delay, mixed, neg slope factor (312.5e-15 Sec counts): -8179
Ch 1 State Delay, pos slope (Sec): 1.06570513e-009
(chan1StateDly[POS_SLOPE]) Ch 1 State Delay, pos slope factor (312.5e-15 Sec counts): -2890
(chan1StateDly[NEG_SLOPE]) Ch 1 State Delay, neg slope factor (312.5e-15 Sec counts): -2890
Ch 1 Timer Delay, mixed, pos slope (Sec): 1.328125e-009
(chan1TimerDlyMixed[POS_SLOPE]) Ch 1 Timer Delay, mixed, pos slope factor (312.5e-15 Sec counts): -10750
(chan1TimerDlyMixed[NEG_SLOPE]) Ch 1 Timer Delay, mixed, neg slope factor (312.5e-15 Sec counts): -10750
iteration: 1
Ch 1 Pattern Delay, pos slope (Sec): -4.00641026e-012
(chan1PattDly[POS_SLOPE]) Ch 1 Pattern Delay, pos slope factor (312.5e-15 Sec counts): -2971
(chan1PattDly[NEG_SLOPE]) Ch 1 Pattern Delay, neg slope factor (312.5e-15 Sec counts): -2971
Ch 1 Pattern Delay, mixed, pos slope delay (Sec): 2.00320513e-012
(chan1PattDlyMixed[POS_SLOPE]) Ch 1 Pattern Delay, mixed, pos slope factor (312.5e-15 Sec counts): -8173
(chan1PattDlyMixed[NEG_SLOPE]) Ch 1 Pattern Delay, mixed, neg slope factor (312.5e-15 Sec counts): -8173
Ch 1 State Delay, pos slope (Sec): 4.00641026e-012
(chan1StateDly[POS_SLOPE]) Ch 1 State Delay, pos slope factor (312.5e-15 Sec counts): -2877
(chan1StateDly[NEG_SLOPE]) Ch 1 State Delay, neg slope factor (312.5e-15 Sec counts): -2877
Ch 1 Timer Delay, mixed, pos slope (Sec): -4.11184211e-012
(chan1TimerDlyMixed[POS_SLOPE]) Ch 1 Timer Delay, mixed, pos slope factor (312.5e-15 Sec counts): -10762
(chan1TimerDlyMixed[NEG_SLOPE]) Ch 1 Timer Delay, mixed, neg slope factor (312.5e-15 Sec counts): -10762
**** CAL PASSED **** Time: 4 seconds
**** Analog Channel Skews ****
Top : 0.251256
Base : -29.497488
skew for channel: 2
adjustment (Sec) =: -3.4375e-010
2Gsa/s factor in 312.5e-15 Sec counts: -1099
skew for channel: 2
adjustment (Sec) =: 4.00641026e-012
2Gsa/s factor in 312.5e-15 Sec counts: -1086
skew for channel: 1
adjustment (Sec) =: -1.30208333e-010
1GSa/s factor in 312.5e-15 Sec counts: -416
skew for channel: 2
adjustment (Sec) =: -9.7265625e-010
1GSa/s factor in 312.5e-15 Sec counts: -3111
skew for channel: 1
adjustment (Sec) =: -1.32211538e-010
1GSa/s factor in 312.5e-15 Sec counts: -838
skew for channel: 2
adjustment (Sec) =: -1.328125e-010
1GSa/s factor in 312.5e-15 Sec counts: -3535
-----< Memory Status >-----
Memory Load 50% -> 53%: 3%
Physical Used (MB) 36.496 -> 38.297: 1.801
Total Calibration Time: 361
Lowest Frame Temp : 19.750000
Highest Frame Temp: 20.750000
Update Cal Header
Factory Cal Mode
Revision : 2
Cal Mode : Factory
Cal Satus : CAL_OK
Firmware Ver : 01.00.2016091301
Cal Temp : 18.750000 C
Cal Date : Tue Sep 27 16:54:25 2016
Cal Duration : 410s
Service Cal Mode
Revision : 1
Cal Mode : Service
Cal Satus : CAL_OK
Cal Temp : 19.500000 C
Cal Date : Tue Sep 27 16:54:24 2016
Cal Duration : 410s
User Cal Mode
Revision : 8
Cal Mode : User
Cal Satus : CAL_OK
Firmware Ver : 01.01.2016092800
Cal Temp : 20.250000 C
Cal Date : Tue Mar 14 20:43:37 2017
Cal Duration : 361s
Calibration PASSED
Board: 0
Inst: CN56396387
Date: Tue Mar 14 20:43:38 2017
Duration: 6.0 minutes
Inst Temp: 20.3 degrees C
Hardware selftest does this
** self test: PASSED : DDR Mem Bus
** self test: PASSED : Acquisition Memory
** self test: PASSED : ADC
** self test: PASSED : MegaZoom SIPO
** self test: PASSED : TrigComp & Mux
** self test: PASSED : FirmwareStatus
