Same thing happens on my U1273A. I did some tests and noticed the Low Pass Filter makes a difference. Here's the data:
Level | LPF On | LPF Off |
500mv | 0.0020A | -0.275A |
750mv | 0.0037A | -1.026A |
1.00v | 0.0064A | -3.306A |
1.25v | 0.0096A | -6.140A |
1.50v | 0.0138A | -9.820A |
1.75v | 0.0180A | -5.420A |
2.00v | 0.0228A | -5.410A |
2.25v | 0.0360A | -6.090A |
2.50v | 0.0440A | -7.480A |
The above readings were taken on my U1273A DMM, reading a 10Mhz 2.5v square wave signal generated from the waveform generator output of an Agilent DSO-X 3034A scope. I also moved the meter as far away from me as the test lead would allow to prevent stray pickup when moving my hand from the scope to the notepad (it did make a difference as noted in some of the earlier posts).
So it seems the LPF has a pretty big impact on the readings. As a bonus, I turned the filter on and took some additional readings. Below 500kHz, the reading barely budged above 0.001A (this is interesting since the manual says the filter's cutoff frequency is 1kHz). From 1MHz to 10Mhz (in 1Mhz steps), the reading went from 0.0016A to 0.0360A, so although the filter seems to have done it's job, it wasn't perfect.
Side Note: Trying to put blame on the software guys (I'm allowed to as I am one), I looked at the firmware updates. Unfortunately, going from v1.95 (mine) to v3.01 (latest) were only these two items:
- Fix incorrect average range in autorange feature
- Keysight rebranding, change identity to Keysight and Keysight Logo
...which is
really funny since my meter is branded Keysight, but boots Agilent!
Maybe the true issue is the meter's dual personality, and ghosts of Agilent-past are wreaking havoc on random readings.
...the mystery continues!
-Frank