Just a remark about the delta markers. When you take a 'difference' in dB's between two values (ie a ratio) the unit disappears. Eg the 'difference' between 23 dBm and 13 dBm is 10 dB. Likewise the difference between -100 dBm/Hz and -120 dBm/Hz is 20 dB.
It seems the Rigol takes the ratio of the powers and then divides by the RBW. For example at 24:23 I read 0 dBm and -67 dBm. the ratio would be -67 dB. Dividing by the RBW results in -87 dB/Hz.
However in my view the only point where this would make sense is when the reference marker is at the peak (as in your case). For any other point except that special case the Siglent way (just dB) is more generic. For example if I would want to compare the noise floor at two points, supposing there is a -10 dB ratio between the two powers, I would rather read -10 dB, than -30 dB/Hz. This latter value would not make any sense to me.