Hi, i can still confirm this issue with my unit (KA3005P), just recently ordered from hotair.pl. (Btw they sadly put stickers on some case screws so I can't open it without losing warranty.)
I tried with 1 and 3S 1W LEDs = 3 and 10V and for both there was a gap of 200mV. This is a >5% error for low voltages. Pretty damn much for a claimed accuracy of 0.5%+20mV!
What a retarded software """feature""".
To summarize misbehavior in few a words:
If in CC-mode V
real approaches V
set to (V
real = V
set - 0,2V) the display switches from displaying V
real to display V
set.
This does not happen for the displayed current in CV-mode though.
Another notsonice fact i discovered:
OVP does not trigger reliably for me: If OVP is enabled and output is open looped it does not trigger (though ofc V
real=V
set). Also if I slowly increase the current through a load until V
real reaches V
set it does not trigger, no matter how much further I increase C
set. It does only trigger if I increase the C
real pretty fast, thus "shoot" in the V
set limit pretty fast.
Don't know how much this matters in practical usage, as OVP won't be used often anyway I guess.
3S=3 in series | CC=constant current | CV=constant voltage | C=current | V=voltage