Author Topic: EEVblog 1593 - 16kV ESD Tester REPAIR - Part 2  (Read 3032 times)
Offline EEVblogTopic starter
« on: January 17, 2024, 10:07:06 pm »
Repairing the Haefely Trench PESD-1600 ESD Tester, plus some testing and discussion of the IEC 61000-4-2 standard.
00:00 - Repair of the Haefely Trench PESD-1600 tester
01:40 - Special snowflake rotary encoder test
03:15 - Laugh at Dave as he solders before looking
05:06 - Nothing worse than a floppy pin
07:09 - Soldering the switches
11:01 - The reassembly nighmare
18:52 - Does it work?
19:53 - Battery pack replacement
22:27 - ESD Testing that's good enough for Australia
The following users thanked this post: T3sl4co1l
Offline chriswebb
Re: EEVblog 1593 - 16kV ESD Tester REPAIR - Part 2
« Reply #1 on: January 18, 2024, 05:54:09 am »
Would it be possible to measure the output without the expensive target? Like with only a dmm or maybe even a scope or spectrum analyzer with proper protection?
Always learning. The greatest part of life is that there will always be more to learn.
« Reply #2 on: January 18, 2024, 06:26:57 am »
Sure, make a 2 ohm load with as much bandwidth and linearilty as possible. Not hard to do, but full charactertisation would be the hard part.
And of course you need a very high bandwidth scope and associated attenuators and cabling.
I might take a crack at it...
« Last Edit: January 18, 2024, 06:51:47 am by EEVblog »
Offline PartialDischarge
« Reply #3 on: January 18, 2024, 01:15:48 pm »
If you are building something for testing, the ground plane in esd setup has to be connected via a 1Mohm (2x470k) resistor back to the esd gun
« Last Edit: January 18, 2024, 01:17:41 pm by PartialDischarge »
« Reply #4 on: January 19, 2024, 05:43:44 am »
It's the horizontal ground plate that need to be connected via 1M to the floor ground plate. Same for the vertical ground plate.