Part 2 in the dumpster Tektronix 4CH 500MHz TDS540S oscilloscope.
Replacing the CRT with an LCD had a few issues...
Ebay listing here: https://www.ebay.com.au/itm/225251897379
The following users thanked this post: thm_w, sean0118, TERRA Operative
Online darkspr1te
  • Country: zm
When i saw the disruption in the initial image i knew it was the dot clock setting, iiyama monitors used to have that option burried in the menus and god forbid you change it to a setting incompatible with the vertical refresh your vga card is outputting right now as you would end up with a corrupted image or heavily overscanned image including the menu which relied in the incoming clock signal to match displayed timing , if you were lucky and had matrox or ati cards they would auto switch or you could change via a command line program which ment sometimes making a boot disc for dos with autoexec.bat running your command to switch vert freq of the vga signal.
ah the memories.
darkspr1te
The following users thanked this post: EEVblog
Offline TERRA Operative
  • Country: jp
  • Voider of warranties
    • Near Far Media Youtube
Nice bodge! :D
Question re. the HDD option in there, is there any software on that HDD by chance? I have some of the TDS software packages but I'm always on the lookout for more. :)
Offline sean0118
I still have Tektronix TDS754A parts available (in Sydney) which maybe be useful for this...    front panel knobs look the same at least...
https://www.eevblog.com/forum/buysellwanted/fs-tektronix-tds754a-parts-only-(sydney-aus)/msg4254691/#msg4254691
Offline bob808
You should have checked the 12Vin circuit of the LCD board. It might have step up/down regulators on board which would be ok with 15Vin. Even a linear regulator wouldn't dissipate much extra heat with another 3V on top. The only issue would be the input caps if they're 16V rated, too close for comfort.
The following users thanked this post: NiHaoMike
Offline nctnico
You should remove the glass from the bezel. That is quite dark and not necessary with an LCD screen fitted.
Offline Andie
Regarding the fan related surprise: Once I replaced the fan on a HP54100A scope, which apparently checks for the ripples in the fan current, presumably caused by commutation. The HP didn't work with modern fans until I injected some ripple from a small mains transformer into the fan current.
Offline thm_w
  • Non-expert
Just sounds like a poor solution to a simple problem.
Either use fan with RPM output (if they had them then), or put a NTC temperature sensor on the board for 5c.
Don't know. It's now packed up ready for shipment.
Offline james_s
Nice bodge! :D
What software was offered? I have the HDD in my TDS784C, I thought it was just for storing captured waveforms and such.
Offline Bud
I have a couple computer fans salvaged from several years back, they have a thermistor built-in, the fan starts ramping up when air temperature hits 37C.
  • Country: jp
Nice bodge! :D
Here's a list of what I know of off hand, there may be a few more but I'd have to dig around to see.
Jitter Analysis - TDSJIT1V2
Disk Drive Analysis - TDSDDM1
Power Analysis - TDSPWR1
Enhanced Printing Functions - TDSPRT1
Communications Pulse Measurements - TDSCPM1
Eye Pattern Measurements - TDSEYE1
Communications Eye Diagram Measurements - TDSCEM1
Processor Specifications Measurements - TDSPSM1
I have Comms Eye Diagram Measurements, Eye Pattern Measurements, Jitter Analysis and Enhanced Printing Functions.
I really want to get my hands on the Power Analysis so I can put my current probe to good use.
Offline james_s
Jitter Analysis and Enhanced Printing could be useful, I doubt the comms eye diagram stuff is good for much anymore. I agree the power analysis sounds interesting. I wonder what Disk Drive Analysis does? That could be useful for retro stuff.
Offline David Hess
  • DavidH
The VGA horizontal and vertical synchronization signals are also 75 ohms, so there should be 5 75 ohm terminations if those are used.  They are actually useful as a fast edge test signal with transition times of 100s of picoseconds.
VGA also supports synchronize on green and if you examine the green output, it usually contains the horizontal and vertical synchronization signals just like composite video.  I suspect the problem was just selecting the correct synchronization source.
Areca RAID controllers do the same thing.  They use a 2 wire fan, and measure the ripple current to determine the RPM or if the fan is even operating.
The following users thanked this post: EEVblog, thm_w, sean0118
Offline strick
Terra,
I have a 580D and a 580C, both with hard drives...
Would you be willing to share the applications you mentioned in your post?  Also, I would assume they have to be activated.  Did TEK do it for you or did you find a way around that?
thanks,
Strick
  • Country: jp
Terra,
thanks,
Strick
PM me your email and I'll send you what I have. I'm not sure what will work on 500 series scopes as I don't have one to test, but you'll soon find out. :)
There's no special activation, just load the particular software on (I removed the HDD from the scope and replaced it with a CF card and adapter, so I copied the software over to the CF card from my PC using a USB card reader).
Once you boot the scope, you can load the software package you want to use from the scope menu system.
You can activate options via GPIB though, like this:
Send the GPIB commands over GPIB after switching the protection switch to unlock and sending the text 'PASSWORD PITBULL' (without the quotation marks) via GPIB.
Option   Name                               NVRAM Addr      Hex Addr         GPIB Command                       Description
  1M     4M acquisition length              0400081F        0000081F         WORDCONSTANT:ATPUT 327686, 1       Extends standard acquisition length from 50,000 points per channel to 130,000 samples on three or four channels, 250,000 samples on two channels, and 500,000 samples on one channel.
  05     Video Trigger                      0400081D        0000081D         WORDCONSTANT:ATPUT 327687, 1       Oscilloscope comes with tools that allow investigation of a range of NTSC, PAL, SECAM, and high definition TV signals.
  13     RS-232C and Centronics interfaces  04000823        00000823         WORDCONSTANT:ATPUT 327688, 1       Add RS-232-C and Centronics interface ports.
  2F     Advanced DSP math                  04000827        00000827         WORDCONSTANT:ATPUT 327689, 1       Add advanced DSP math features such as FFT, integration, and differentiation.
  1F     Floppy disk drive                  04000825        00000825         WORDCONSTANT:ATPUT 327690, 1       Add 3.5 inch, 1.44 MByte floppy disc drive.
  2C     Communication Signal Analyzer      0400082B        0000082B         WORDCONSTANT:ATPUT 327692, 1       Oscilloscope comes configured for communications signal triggering and mask testing.
  3C     P6701B with system calibration     0400082D        0000082D         WORDCONSTANT:ATPUT 327693, 1       Oscilloscope comes with a P6701B and calibrated short-wavelength optical reference receiver on channel 1.
  4C     P6703B with system calibration     0400082F        0000082F         WORDCONSTANT:ATPUT 327694, 1       Oscilloscope comes with a P6703B and calibrated long-wavelength optical reference receiver on channel 1.
  2M     8M acquisition length              04000831        00000831         WORDCONSTANT:ATPUT 327695, 1       TDS 520D, 724D: To 2M samples on two channels and 4M on one channel. TDS 540D, 580D, 714L, 754D, 784D, 794D: To 2M samples on three or four channels, 4M on two channels, and 8M samples on one channel (Option includes a hard disk and is only available for the models listed above.)
Note, some options need hardware to be installed (05, 13, 1F, 3C, 4C, 2M). Option will simply not appear if the hardware is not present.
Offline strick
Thank you Terra!  I'll send you a PM with my email address.
Strick
Offline strick
I can verify that the TDS580D with a hard drive and 2M does run the TDS784D software. This included Comms Eye Diagram Measurements, Eye Pattern Measurements, Jitter Analysis and Enhanced Printing Functions.
 I found that I had to install a PCMCIA to CompactFlash adapter in order to put the CompactFlash in my laptop.  The regular PCMCIA hard drives are double slotted and wouldn't fit in the laptop.  I was able to copy TDSCEM1 (Comms Eye) via a floppy to the regular hard drive on the 580D and it worked, but the rest were too large and had multiple levels.  These just failed to properly write to the hard drive.
Strick
Offline Trax
  • Country: at
I just got an old TDS 784D, it has a bad display, you can somewhat see something but only a bit of the waveforms, the text is not readable.
Is this failure more of the CRT known to be reparable, or do I need to swap it for an LCD like in the video?
EDIT: i noticed that when i switch the display to hard-copy preview it kind of gets inverted and then everything is readable just on a white background.
Can it be that I just need to tweak some pot to fix that properly?
@Terra would your files work for my model as well and if so could you please share.
« Last Edit: February 21, 2023, 08:11:31 pm by Trax »
  • Country: jp
Can you provide a photo?
Offline Trax
  • Country: at
Here you go: https://ibb.co/5vmXcnn
The "Hardcopy Preview" comes out with messed up colors on the crammer but for the human eye its just fine, and looks completely white with color lines and black grid and text, etc...
must be due to the faster shutter speed auto exposure sets in combination with how the CRT creates color its not your typical shadow mask mut somethign called NuColor, anyhow...
The Normal and Bold images look on cammera considerably better then to the human eye, due to toe longer exposure you can see the UI elements on the photo quite easy, but to the human eye thay are almost unreadable in booth modes.
  • Country: jp
Yeah, to get good photos you need to slow the shutter speed or similar to capture the whole 'colour cycle' as it flips between red, green and blue with the LCD shutter. Cameras are often faster than the CRT assembly in these units.
Looking at those images, it looks like you may need to start troubleshooting with a run through the CRT contrast/brightness calibration.
Check out the service manual at tekwiki, give it a tweak and let us know how you go.
https://w140.com/tekwiki/wiki/TDS784
Also, do a full factory reset on the unit too, just to make sure there's no funky setting somewhere that is messing things up.
<shift>, <Display/Utility> then you should see 'Tek Secure Erase Memory' highlighted on the bottom of the screen. press the lowest soft-button to the right of the screen to select 'OK Erase Setup & Ref Memory' to reset the machine.
If you don't see that by default after entering the Utility menu, press the left most soft-button along the bottom of the CRT to scroll through the options until you get to the 'Config' option.
Now select 'Tek Secure Erase Memory' and press the 'OK Erase Setup & Ref Memory' as mentioned above to reset the machine.
« Last Edit: February 22, 2023, 08:08:09 am by TERRA Operative »