Note: You need to log in to see the links in this post. If you do NOT see any links, create a forum user and log in.Background:
The hack itself is simply updating the serial number and the model number but does not work anymore on Rigols with firmwares higher than 02.02.SP2, this guide has the files needed to downgrade to 02.02.SP2 , change the serial number and the model number using USB and then to upgrade again.
The downgrade is currently only possible for scopes who come with an firmware up to 00.02.05.02.00
See below , in step 2) is shown how to check your firmware versionLinux users should be able to use
Python with USB, .
or
echo to the USB mount
WARNING:
While this hack was successfully done by a lot of people there is always something that can go wrong.
It IS possible to brick your Rigol when downgrading / upgrading firmware .
So do this at you own risk.Other scopes who are know to be hackable:
the Tekway DST1102B Steps to do:
1) Take the DS1052E out of the box Check if everything works as mentioned in chapter 2 of the Service Guide DS1000E.pdf
Yes, reading a manual before everything fails is not a bad idea sometimes...
2) find out your CURRENT firmware version.To find you hardware revision chose on your scope the "utilities" button, , the lower button until 3/3, choose "system info". This will give the current firmware (not the full version string) and some other info.
example:
Once in this screen press button "ch1" twice, then "ch2" twice then "math" once.
The hardware version will also appear next to "Hardversion". Note if this is 57 (or lower) or 58 .
Be aware !!!!!!! Do NOT press any key except RUN/STOP to exit while in this mode or you can mess up your scope.
Again, do NOT press any key except RUN/STOP to exit while in this mode.example:
Do NOT press any key except RUN/STOP to exit while in this mode or you can mess up your scope.some examples of the firmware versions you might see:
00.02.02.02.00 , shown as 00.02.02 SP2 in the previous screen
00.02.04.00.01 and 00.02.04.00.03 , both shown as 00.02.04 in the previous screen
00.02.04.01.02 , shown as 00.02.04 SP1 in the previous screen
00.02.05.00.00 , shown as 00.02.05 in the previous screen
00.02.05.01.00 and 00.02.05.01.02, both shown as 00.02.05 SP1 in the previous screen
3)a) Get the needed files and save this in a temp folder ( C:\temp ):* Download a tool like
hashcalc to generate MD5 sums of files.
*
visa462runtime.exe :
the visa462runtime.exe is available here (54Mb) , this if for Windows 2000 up to Windows 7, see the readme.html on that website.
*
Ultrascope for DS1000E Series.exe : . Make sure you use the "(VISA version)" one if your scope is older and you had this already installed.
When in doubt deinstall and re-install .
I uploaded the one I used
to rapidshare here (12Mb).
This is the version that came on the cdrom with my scope which says version "00.01.07" in the help - about box.
There is also one
on Rigol's website*
DS1052_to_DS1102.zip : the DS1052_to_DS1102.zip (11Mb) is available here on
Rapidshare or
Fileserve * download
rigupgr.zip from
shafri's post here.
2)b) Install everything you need on your windows pc.Do NOT connect the Scope yet* Install a tool like
hashcalc to generate MD5 sums of files
* Run
visa462runtime.exe to install the usb driver/ VISA runtime.
* Run
Ultrascope for DS1000E Series.exe to install Ultrascope. (This will also install some dll files who are needed for step 5 )
* unzip the
DS1052_to_DS1102.zip in c:\temp
* unzip
rigupgr.zip in c:\temp
I suggest to use 7zip it's free and works great. After the unzip you should have now these directory's in c:\temp:
You need to reboot your pc nowDo NOT connect the Scope yet3) Downgrade or upgrade to 00.02.02 SP2 if needed.Do NOT connect the Scope yet3)a) your firmware is 00.02.02 SP2 (00.02.02.02.00 ).go to point 4
3)b) your firmware is LOWER than 00.02.02 SP2 (00.02.02.02.00 ).If your current firmware is LOWER you need to upgrade to 00.02.02 SP2 to be able to use the USB connection method (point 4),
you should be able to use the "hack" itself without upgrading ( using a RS232 cable) but the USB connection method ( using "VISA" drivers ) will not work .
Easy steps to update the firmware to 00.02.02 SP2:
* empty an usb stick
* check if your usb stick works on the Rigol by plugging it in and saving some screenshots or so - see "to store and recall" in the user manual.
* unplug it again from the scope and plug it in your computer, check files if needed
* delete all files and folders from the usb stick
* copy the DS1000
EUpdate.RGL from the "C:\temp\DS1052_to_DS1102\02.02.SP2_original" directory to the empty USB stick, there should be no folder and only only the RGL file on the usb stick
* "eject" the USB stick in Windows Explorer to make sure windows saved everything, plug it in again.
Double check your USB stick is working on the Rigol and the file is saved properly, seen at least one person bricked his scope due a bad downgrade of the firmware discussion here Check the MD5 sum of the DS1000EUpdate.RGL file on your usb stick.
the MD5 sum for this file should be 272086b2037231c62446617436544a77 * unplug again
* plug the USB stick in the rigol and follow the Rigol menu "upgrade" instructions
If the Rigol does not detect the firmware use another USB stick or double check the file used.* switch off and on the rigol when asked to do so
Check your firmware version of the DS1052E (Utilities button, lower button until 3/3, choose system info) it should list 00.02.02 SP2 .
goto point 4)
3)c) your firmware is HIGHER than 00.02.02 SP2 ((00.02.02.02.00 ) but LOWER than 00.02.05 ( 00.02.05.00.00 )example:
you
need to use a "patched" 00.02.02 SP2 firmware file in the "02.02.SP2_patched_to_02.04.02" directory
Easy steps to downgrade the firmware to 00.02.02 SP2 using the patched file:
* empty an usb stick
* check if your usb stick works on the rigol by plugging it in and saving some screenshots or so - see "to store and recall" in the user manual.
* unplug it again from the scope and plug it in your computer, check files if needed
* delete all files and folders from the usb stick
* copy the DS1000
EUpdate.RGL from the "C:\temp\DS1052_to_DS1102\02.02.SP2_patched_to_02.04.02" directory to the empty USB stick, there should be no folder and only only the RGL file on the usb stick
* "eject" the USB stick in Windows Explorer to make sure windows saved everything, plug it in again.
Double check your USB stick is working on the Rigol and the file is saved properly, seen at least one person bricked his scope due a bad downgrade of the firmware discussion here check the MD5 sum of the DS1000EUpdate.RGL file on your usb stick.
the MD5 sum for this file should be 8cd4e61ce6128b55ab18fc83fa756e34 * unplug again
* plug the USB stick in the rigol and follow the Rigol menu "upgrade" instructions
If the Rigol does not detect the firmware use another USB stick or double check the file used.* switch off and on the rigol when asked to do so.
Check your firmware version of the DS1052E (Utilities button, lower button until 3/3, choose system info) it should list 00.02.02 SP2 .
(this is not a typo - the file was changed to "appear" to be a higher FW version when checked by the rigol's current firmware, but after the flash it does report as 00.02.02 SP2 - the actual version of the firmware
itself is NOT changed since the actual firmware is take from a different part of the file, which is untouched )
goto point 4)
3)c) your firmware is 00.02.05 ( 00.02.05.00.00 ) or 00.02.05 SP1 (00.02.05.01.00) example:
you
need to use a "patched" 00.02.02 SP2 firmware file in the "02.02.SP2_patched_to_02.05.01.02" directory
Easy steps to downgrade the firmware to 00.02.02 SP2 using the patched file:
* empty an usb stick
* check if your usb stick works on the rigol by plugging it in and saving some screenshots or so - see "to store and recall" in the user manual.
* unplug it again from the scope and plug it in your computer, check files if needed
* delete all files and folders from the usb stick
* copy the DS1000
EUpdate.RGL from the "C:\temp\DS1052_to_DS1102\02.02.SP2_patched_to_02.05.01.02" directory to the empty USB stick, there should be no folder and only only the RGL file on the usb stick
* "eject" the USB stick in Windows Explorer to make sure windows saved everything, plug it in again.
Double check your USB stick is working on the Rigol and the file is saved properly, seen at least one person bricked his scope due a bad downgrade of the firmware discussion here check the MD5 sum of the DS1000EUpdate.RGL file on your usb stick.
the MD5 sum for this file should be 7cc0aac2e39f6a677dd0296344e33954 * unplug again
* plug the USB stick in the rigol and follow the Rigol menu "upgrade" instructions
If the Rigol does not detect the firmware use another USB stick or double check the file used.* switch off and on the rigol when asked to do so.
Check your firmware version of the DS1052E (Utilities button, lower button until 3/3, choose system info) it should list 00.02.02 SP2 .
(this is not a typo - the file was changed to "appear" to be a higher FW version when checked by the rigol's current firmware, but after the flash it does report as 00.02.02 SP2 - the actual version of the firmware
itself is NOT changed since the actual firmware is take from a different part of the file, which is untouched )
goto point 4)
3)d) your firmware is 00.02.05 SP1 ( 00.02.05.01.02 ) or higher andrewK just provided a firmware
see this post , which should work for 00.02.05.01.0
2 scopes.
The steps should be just the same as point 3c)
After the scope is on 02.02 SP1goto point 4)
For 00.02.05.
02.00 scopes see
this post
4) connect the scope to your PC and test the USB connectionthe ultrascope step is not needed as such for the "hack" it's simply to see if your usb connection works fine.Start the DS1052E and connect the scope with an USB cable to your PC, a popup should ask you install drivers, choose "don't search for updated drivers" and let windows install the drivers.
If Windows mentions that the "installed hardware does not function properly" then, disconnect the scope, reboot, connect the scope again.
Start ultrascope, click on the little icon "connect" icon in the upper left corner (right from the floppy) to connect to the scope, in the left panel double click on "virtual console", click on "Run" under the "control" and play around to see if it works.
If it does then fine, exit the "virtual console" and then ultrascope and go to step 5.
It might be also a good idea to stop and start the scope again - mine sometimes "hangs" when exiting ultrascope.
If it does not work then there is a problem with the usb drivers or maybe you installed an old Ultrascope, in any case do not go further until Ultrascope works on this PC.
5) Change the DS1052E to DS1102Eyes, we are finally there
5a) automaticallyForum
user shafri has provided a windows "rigupgr.exe" program that basically does the steps in point 5b) automatically .
The conditions are still the same, this will ONLY work on 00.02.02 SP2 scopes.* Connect the rigol with the usb cable to your pc if not yet done
a) start rigupgr.exe from c:\temp\rigupgr (or where you extracted rigupgr.zip)
b) Click "Check Rigol Model & Serial" button, it will show you the *IDN? result and your FW version. (it will automatically fill the new model and serial number for upgrade process)
c) Click "Upgrade Rigol" button to start the upgrade. Wait until it ask you to restart your rigol and click OK.
Dont do anything in between, this might take a while.
restart the scope and goto point 6)
5b) Manually* Connect the rigol with the usb cable to your pc if not yet done
* Start demoIDN.exe from the "C:\temp\DS1052_to_DS1102\usbtool" folder
* select the scope in the first "Visa" drop down list (there should only be one line), then select "*IDN?" in the bottom drop down list and hit "send and read"
It will give:
Rigol Technologies,DS1052E,DS1ED888888888,00.02.02.02.00
Instead of 888888888 you will see
your serial number, copy paste the serial number into notepad, you need it later.
The Scope MUST have 00.02.02.02.00 as firmware to do the hack.* In notepad prepare 2 lines:
:INFO:MODEL DS1102E
:INFO:SERIAL DS1EB888888888
where 888888888 is
your serial number, note the difference between the original DS1E
D and the new DS1E
B.
* Next in the bottom select box of demoIDN.exe copy paste:
:INFO:MODEL DS1102E
and click on the send and read button - no output is given, that's normal
* After that , in the bottom select box of demoIDN.exe copy paste:
:INFO:SERIAL DS1EB888888888
where 888888888 is your serial number you prepared before, make sure you use DS1E
B not DS1E
Dand click on the send and read button - no output is given, that's normal
Now exit demoIDN.exe and switch off the rigol and then switch it on again.Start demoIDN.exe again, select the scope in the first drop down list, select *IDN? in the bottom and hit send and read
it will give
Rigol Technologies,DS1102E,DS1EB888888888,00.02.02.02.00
where 888888888 is your serial number
and your DS1052E scope is now a DS1102E....
got step 6
6) Update your firmware after the change.Get your hardware version, that is the "Hardversion" found in step 2 of the guide.
HardWare version "57" (or lower) scopes MAY stay on 00.02.02.02.00 but there no reason why you should NOT go to 00.02.04.01.02 (00.02.04 SP1)
HardWare version "58" scopes NEED to go to 00.02.04.01.02 (00.02.04 SP1) other wise there are calibration issues
HardWare version "59" scopes are reported to exist , not clear yet what Firmware version the NEED, suggest to try with 00.02.04 SP1 first and report
from page 33 onwardsThere seams currently no reason to go to 00.02.05.01.00 for HardWare version "58" (or lower) scopes , some people even reported issues with 00.02.05.01.00 firwmare that do not occur in 00.02.04 SP1,
like this oneThe 00.02.04.01.02 (00.02.04 SP1) firmware can be found in the "C:\temp\DS1052_to_DS1102\02.04 SP1_original" directory.
Again, make sure your is file correct
on the USB stick, you can use a tool like
hashcalc to check the MD5 sum of the DS1000EUpdate.RGL 00.02.04.01.02 (00.02.04 SP1) update file
on your usb stick.
The MD5 sum for this 00.02.04.01.02 (00.02.04 SP1) DS1000
EUpdate.RGL file should be
bcf73565352391935f3a9651d30776ee If you really want you can go to 00.02.05.01.00 ( 02.05 SP1 )
The 00.02.05.01.00 ( 02.05 SP1 ) firmware can be found in the "C:\temp\DS1052_to_DS1102\02.05.01.00_original" directory .
The MD5 sum for this 00.02.05.01.00 ( 02.05 SP1 ) DS1000
EUpdate.RGL file should be
839651c74a948cea1d8cccc21af9fc8a
7) After you are on your final firmware , let the scope run for 30 minutes to warm up and perform a self calibration as the very last step.
And do not forget to delete the *.RGL file of the usb stick after doing an update of the firmware
If you want even more MHz , help out in the
DS1052E Up to 150Mhz Thread
OR you want less noisy fan:
Rigol 1052e Fan Replacement enjoy
Footnotes:* To change an DS1052D (the model with Logic Analyser) into an 1102D the steps should be just the same, the serial number however should be changed from DS1E
Cxxxxxx (1052D model) to DS1E
Axxxxxx (1102D model)
See here.
* The pfd's in attached rigolpdf.zip are the DS1000E_programming guide.pdf from dec 2009, User’s Guide DS1000E.pdf from July 2009 and Service Guide DS1000E.pdf from may 2009.