I hope you got the special Pink edition
I had requested a customized pink version ... however the pink of channel 2 is very nice
Siglent manual that comes with the unit doesn't tell you a lot of things.....
About the only way you can screw it up is failed firmware update.
oh well, meanwhile, I start learning what the manual says, all new things for me ..
For the firmware I shouldn't risk anything, I have the latest version
1x might be useful when the voltage is very low and the frequency is also low. None of these apply to me so I would rather have fixed 10x probes. In fact, I bought a pair of these from Probe Master for my Analog Discovery 2. It has very limited allowable input voltage and it is also expensive. I want to protect it as best I can.
I would almost recommend a dab of epoxy on the switch to keep it in the 10x position forever.
very clear, i will only use 10x. In fact it would be better to have 10x fixed probes. A drop of hot glue could be useful on the swich
The ground spring is useful for high frequency signals that would get distorted because of using the ground clip. Detaching the ground cable and clip and using the spring instead makes the probe behave like a "better quality cable".
Well, if you want to probe something on a connector it is really convenient instead of playing games with the ground clip and trying to keep the probe tip in the right position.
Also, some circuits may have purpose made test points that can have a connector. Or at least those test points can be designed to make probing with a ground spring easier.
ok for the use of the spring: however, using the spring, it is necessary to have the ground point and the measuring point very close together, otherwise it is impossible to place the tip and spring..
So in practice that bnc connector, for example if you want to measure another bnc connector with the probe, safely and easily.
The major difference between 1x and 10x is that 10x has much higher impedance. So it will divert less, and so it will affect your circuit less. When you have a small and weak signal, if you use 1x, so much of the signal get sucked into your scope that circuit may not work. Use 10x to lessen this effect.
Also, 10x mode has higher frequency bandwidth. Our probes are designed up to 200MHz. But that spec only applies to 10x mode. In 1x, it's less. (I don't remember what the limit was) So most of us stay in 10x and only cautiously and when necessary, use 1x.
all roads lead to the 10x probe ... they shouldn't even have three meters with 1x swich, I would have preferred 10x fixed probes now that I understand how it works
If you remove the end of your probe with the spring hook from your probe you will be able to push the small end of this adapter on the pin sticking out the end of the probe. The large end of the adapter is like a BNC cable connecter (without the outer locking shell) to allow you to just push it on a connector without locking it in place and will allow you to connect your probe to the output of a lot of test equipment. If you have a function generator with a BNC output connector you can use this adapter to connect your scope probe to the generator output and still use the 10X setting. You could use a BNC to BNC cable between the two instead of your probe but it would only be straight through at 1X.
ok, now I understand what it's for, in fact I tried to put it on the tip of the probe and it's perfect. In practice it is used to measure with the oscilloscope BNC outputs positioned on other instruments or on circuits, to see the wave generated for example.
thank you