Are there any gotchas when putting a function generator in series with a DC supply?
Any risk if the current draw is significant?
I'm working on a power supply and trying to simulate common mode noise/mains noise/etc, such as what might be seen after noisy AC rectification.
Is there a better way to go about this?
I found the following document (
http://cp.literature.agilent.com/litweb/pdf/5990-3466EN.pdf ):
If you want to produce a signal riding on top of a DC offset,
but you require a DC offset larger than your function generator can
deliver, you can put a DC power supply in series with the function
generator output, as shown in Figure 3 above. Using this technique,
you will have the full bandwidth capabilities of the function
generator available to you in addition to having fl exibility in the
DC level provided by the power supply. However, this technique
has a few important limitations. The output of the function generator
may or may not be internally isolated (fl oating) from earth
ground. When it is isolated, there is a specifi cation indicating the
maximum amount of voltage off of earth ground that you can fl oat
the output. For example, the Agilent 33220A’s output is isolated
from earth ground and you can fl oat the output off of ground up to
+/- 42 V. This means that if you put a DC power supply in series
with the function generator output, the DC offset must be less than
+/- 42 V. If the function generator output is internally connected
to earth ground, then the power supply output voltage must be
isolated from earth ground (unless you plan to connect the earth
grounded nodes together). The vast majority of Agilent power
supplies have outputs that are isolated from earth
This makes it sound like no big deal, but I'm a bit hesitant to put my DC supply in series with my function gen.
Edit: Using DG1032Z and DP832.