Do you have a device anywhere in the house which has a NTSC composite video output,such as a Digital Set Top Box,DVD player,etc.
The Field rate in NTSC is very closely controlled at 60Hz,so there is a second check for your period.
The line rate is approx 15.750 KHz *,which gives you a line duration of 63.49 microseconds.
The line sync pulse width is around 4.7 uS.--Please forgive me for not knowing this precisely,as I live in a PAL country!
If you terminate your Oscilloscope in 75 Ohms,& connect it to the composite NTSC video source with a 75 Ohm coax
cable ( for a short length,you can get away with a 50 Ohm cable),you should be able to get a fair check of your 'scopes performance using known signals.
These signals will not test the high frequency limits of your scope,but will let you see if there is anything basically wrong with it.
*I say approx,because they fiddled it slightly when NTSC colour was devised.
It was exactly that figure for the original B&W System.