My second unit from this supplier is already in the post to me....nice fast despatch
I have already identified several multimeters that I own that are less accurate than my Fluke's. OK they are not miles off spec but they could do with either calibration or marking as not 'Fluke' precise instruments. Thankfully I have my Fluke and Advantest bench meters if I need really accurate readings. It was interesting to see the variation in readings across some of my collection of multimeters though. My AVO Megger units were as accurate as my Fluke 87 III. My Wavetek DM23XT, less so.
My Voltage reference is stated as providing the following voltages at 21 degrees C :
2.50093V
5.00165V
7.50336V
10.00360V
My old Uni-T UT39A gave the following readings:
2.48V
4.96V
7.45V
9.93V
Another old meter...the Wavetek DM23XT manual ranging:
2.50V
5.01V
7.52V
10.02V
Old Wavetek DM73A Probe meter:
2.504V
4.99V
7.49V
9.99V
My Fluke 87 III readings:
3.5 Digit mode:
2.500V
5.00V
7.50V
10.00V
4.5 Digit mode:
2.501V
5.002V
7.503V
10.004V
A simple example of the difference between a relatively old Fluke 87 III and a UNI-T meter.... One remains accurate, the other not so accurate
....but was it ever 'Fluke' accurate