Hello!
Thank everybody for the answers and feedbacks.
I have made four pictures:
They are all made with CH1 only (CH2 is disabled) and nothing attached to CH1 (nor CH2).
This has been done after about 30min of operation and a "calibration" procedure, including "quick cal = ON".
1)
Noise with Vertical Res = 0.5mv and Horizontal = 2ms. The statistics shows that the average is 93uV.
2)
Noise with Vertical Res = 0.5mv and Horizontal = 2us. The statistics shows that the average is 95uV, which agrees with (1) !
3)
1Mhz 0.5mV Rms source, Res = 0.5mv and Horizontal = 2us. The signal is quite noise, the average is 60uV.
4)
1Mhz 0.5mV Rms source, Res = 0.5mv and Horizontal = 2us. Here I have set CH1 bandwidth to 20Mhz and Trigger "Reject HF". The signal is much less noisy. Average is 31uV.
So in all scenarios, I get a deviation from the ideal "0V average" of some 100s of uV (or less).
I am very happy with the scope overall. I just want to be sure if this type of deviation from the ideal 0V is within spec.
From what I understood from your comments/help, it looks like it is working within specifications.
Is the fan working.
yes.
EDIT: The 0.5mV rms signal has been generated by a Siglet SDG 1032X, which is working fine.