I've been testing an HP 4436A 600Ω balanced attenuator but there's something about the balanced connection that I just can't get right.
The setup described in the manual is simple: A signal generator is configured to 600Ω output and hooked to the attenuator HIGH and LOW input, the GROUND input goes to cable shield. The HIGH and LOW outputs are then connected to a multimeter to check how much voltage is being dropped, lastly 2 * 300Ω shunts are placed to match the high impedance of the voltmeter to the 600Ω output impedance of the attenuator.
The problem is I reproduced this setup but my readings are off. The attenuation is adjustable and will work allrightish from 0dB down to around -80dB but from -80dB to -120dB the voltage doesn't drop at all. This is my implementation of the setup (
I swapped the DMM for a lockin amplifier with floating differential inputs because of the very high attenuations).
Initialy I suspected a hardware error but the attenuator is fine, no resistor has drifted out of tolerance, all switches make clean connections, everything is up to spec. To drive the point home the manual (
attached) specifies the attenuator can be set up with an unbalanced connection so I checked and sure enough when configured as the image below it works flawlessly down to -120dB
Hooking the generator output to attenuator LOW (instead of HIGH like in the picture) will also work equally well so both T attenuators are working perfeclty. This has me thinking it's got to be user error, there has to be something about the balanced setup that I'm misunderstanding
.
Thank you so much.