Well.
According to the instructions, I was able to better understand how this excellent machine works.
1 - The selector must be in the "warm up" position. A negative number will start to run fast (-00.83000) and then, around one hour, this incremental number will run slow, very slow (-00.94000) until you see this "race" becomes very slow, giving the impression of not running anymore. I understand that the device is ready!
2 - In O2 position, it´s possible to adjust the potentiometer until you see (+000000) alternating with (-000000), for me, the voltage zero is done.
3 - In O1 position, it´s not easy to set the zero, but something near of it, by the speed of variation between (+000065) and (-000050).
4 - Checking again O2, need a small adjustment again.
5 - Checking again O1, another small adjustement back to 4 until you a balance between the two settings. At this point, nothing is connected to the inputs.
6 - Calibration position, I started with a voltage of 2.5V and saw in the display: (+1.020000), so I turned the dial from 1 to 10, nothing changed. I understood that for this calibration, it is necessary, through the back connectors, the voltage of 1.0V.
7 - I put 1V and I understood that the selector was just a fine adjustment, for example: If the selector is in position 5, adjust the potentiomentro to exact 1.000000v, in position 6 it will be 1000100v, in position 7 will be 1,000200ve in position 4 will be 0.000900v. It is as if the selector was a fine adjunct, some actuation of the amplifier to hit the input voltage.
I have two challenges ahead: I need to make a connector with probe cables, I thought of making a board that fit perfectly in the connectors and with a selector switch for 1kV and another to connect U with I and U * with I *. The other challenge is to make a stable reference of 1.000000v.
Attached, some images.
Spasibo!!