No, em132 could be correct. It is possible that the UT61E could meet the higher CAT ratings with the right fuses in place but the bean counters didn't want to raise the cost of the meter to do it. That is also assuming that the different fuses do cost differently enough to make this decision.
But.....
My UT61E has a manufacture year of 2011
The requirements for measuring instruments from the ISA 2004 :
"16.2 Multifunction meters and similar equipment
Multifunction meters and similar equipment shall not cause a HAZARD in any possible combination
of RATED input voltages, and settings of function and range controls. Possible HAZARDS include
electric shock, fire, arcing and explosion.
Conformity is checked by the following test.
The maximum RATED voltage specified for any function is applied to each pair of TERMINALS in
turn, in every combination of function and range controls.The test source connected to the
equipment measuring TERMINALS during this test is limited to3.6 kVA for measurement category I
or measurement category II. For measurement category III or measurement category IV, the test
circuit has to be capable of delivering 30kVA.
During and after the tests, no HAZARD shall arise.
Multifunction meters and similar equipment are to be tested by changing the Function/Range Selector to all
possible settings while connected to the maximum rated source.”
It is obvious that the UT61E cannot meet those specs and never could.
There are requirements for markings too but I need to dig those out too. So maybe it is a fact and not just my opinion.
reference thread:
https://www.eevblog.com/forum/projects/cat-ratings-and-interpretation/EDIT:
OK, I hijacked this thread a bit in discussing the failings of earlier Uni-Ts. My point has been to say that I would really like to see these new Conrad meters and their interiors. I would be very happy if they are really built correctly. If they are, and to the point of the OP's question, then they might be great meters for their price.